i dont measure the supply voltage, when meassure the output voltage at pin 6 the one power supply decrease from 15V this is the problem maybe i must isolated then ground of power generator and osciloscope ?
deacrease the -V near zero ( short out)Which supply? V+ or V-
WHen does it decrease and how much?
Then your drawing is incorrect. One of your grounds is shorting the negative supply.deacrease the -V near zero ( short out)
how is the correct setup of expirement ?Then your drawing is incorrect. One of your grounds is shorting the negative supply.
I have seen that type of issue before with some scopes. What brand scope are you using?i dont measure the supply voltage, when meassure the output voltage at pin 6 the one power supply decrease from 15V this is the problem maybe i must isolated then ground of power generator and osciloscope ?
Then you are using a 15 volt supply whose Negative Terminal is internally connected to Ground. If you now try to ground the Positive terminal, you are effectively shorting the power supply.deacrease the -V near zero ( short out)
It is correct as drawn. What I am saying is that your actual wiring is different than drawing. The post above suggests one way that might be true, i.e. there is a ground connection you are not aware of.how is the correct setup of expirement ?
I have seen that type of issue before with some scopes. What brand scope are you using?
Using a capacitor in series with the scope probe might be all that's needed.
That answer makes no sense at all!!i dont measure the supply voltage, when meassure the output voltage at pin 6 the one power supply decrease from 15V this is the problem maybe i must isolated then ground of power generator and osciloscope ?
If you don't measure it, how do you know that it has changed?i dont measure the supply voltage, when meassure the output voltage at pin 6 the one power supply decrease from 15V this is the problem
If you are doing a one-terminal measurement, which is exactly what the TS is describing, the result will be exactly as accurate as any other single point voltage measurement would be.If you don't measure it, how do you know that it has changed?
the supply voltage is ok... the problem is the current the current in both V+ and V- is not the same !If you are doing a one-terminal measurement, which is exactly what the TS is describing, the result will be exactly as accurate as any other single point voltage measurement would be.
And if the TS is not willing or able to tell us the reference point for the claim that the one power supply voltage has decreased then there is no reason to continue this thread. Mb2