That's correct.This is my understanding anyway, if I'm wrong I'd dearly love to understand why!
So how is that different from what I said?
That's correct.This is my understanding anyway, if I'm wrong I'd dearly love to understand why!
I saidThat's correct.
So how is that different from what I said?
Okay, I see the misunderstanding.that one can get an electric shock if both the DUT and the scope are isolated,
Since the isolated secondary can't generate any voltage to the safety earth ground, that would seem to be true.So as long as the secondary winding has no connection to safety ground then having a secondary safety earth ground is allowed?
In my mind this would eliminate a respected contributor of AAC.I want to get an isolation transformer, in my mind this eliminates the need to ground connected appliances and or if they are grounded (as in their plug has a ground pin), there's no path to ground if I accidentally touch something I shouldnt
I do not think your loved ones would find solace in your untimely demise, because of your beliefs.They claim you need both an isolation transformer with the refereneced ground and a surge suppressor to do any real good. I believe them
The isolated transformer's outlet or wall outlet?the outlet you work from should have GFCI.
That article from Tektronix has to be trusted, but I do not understand something and this is what's coming up time and again as we discuss this, here a snapshot from that article:Just be careful. Isolation transformers for utility line operated test and measurements are banned at many industrial sites. Your instrument might be floating but you as the operator are not. The operator panel and knobs and buttons in the test instrument are not usually designed for large differentials between chassis 'ground' and the operator.
https://www.newark.com/pdfs/techarticles/tektronix/FFM.pdf
I expect it can only apply to the wall outlet, because the GFCI is sensitive to leakage currents (to ground) and relies on a ground referenced outlet, I don't think a GFCI would do anything on the secondary side, no leakage (to ground) is possible, there's no way that the current into one side of the secondary can ever differ from the current emerging from the other side.The isolated transformer's outlet or wall outlet?
Wall outlet, which will protect from any fault to ground in the isolation system.The isolated transformer's outlet or wall outlet?
No, the GFCI requires no ground connection to work.the GFCI is sensitive to leakage currents (to ground) and relies on a ground referenced outlet
That's what I meant, it detects (by measuring current delta) leakage (to ground), leakage has to be a possibility so requires a ground referenced outlet, yes the GFCI itself cares not about ground, but unless the outlet's neutral is grounded no leakage is possible so the device serves no purpose, there can never be a difference in current unless the supply its attached to is a grounded supply.No, the GFCI requires no ground connection to work.
It measures the difference in current between hot and neutral, and trips if there's more than a few mA difference, indicating current to ground.
Other than "Touch an OPEN Neutral and Ground - current can flow - you can get a shock" I agree, this is exactly how I perceive it. I think in the Tektronix warning they should have pointed out that their UUT was assumed to be connect to the mains (not isolated) and only the scope was isolated, that would make their explanation consistent.Lots of back and forth here. So here's my answer to HOW an isolation transformer works. As to "When" - That's been covered quite well.
For clarification: Ground is referring to EARTH GROUND.
Below:
Primary side:
Touch Line and Neutral - current flows - you get a shock.
Touch Line and Ground - current flows - you get a shock.
Touch an OPEN Neutral and Ground - current can flow - you can get a shock.
Secondary side:
Touch either leg of the secondary and ground - No current flows - No Shock!
Whether an appliance is grounded or not - if it's isolated from the mains power (from the pole) there is no shock potential for you or a blown Circuit Under Test. Even if you connect a scope and touch the wrong thing - my understanding is there's no blown scope or circuit.
View attachment 231562
OK that makes sense, always a risk in house wiring too unless one is careful.An "Open" neutral can be a loose wire nut, loose screw terminal, broken wire anywhere before it gets back to the panel. Touch before the open circuit and you can get a shock. The illustration below is of a lightbulb that SHOULD be on. But because of a bad (open) neutral there is no return path, no current flow. Touch the lightbulb and current will flow but the filament will act as a resistor and you will get quite a good shocking. Potentially terminal. Hence; you CAN get a shock. But it depends on other factors; where you touch, what the load is, etc.
View attachment 231568