Ohm's Law.Why is that if the amplifier input impedance is specified as >100 MegOhm the amplifier will be very sensitive to radiated mains interference if its inputs are not coupled to a much lower impedance?? Please elaborate. Thank you.
It may not use that. I don't know. The following is the spec of the g.USBamp (worth $17700 but I got only low price used but still I can't take chances to build 7.2V and bring down to 5V unless it is commercially available). It appears to use ADCs to cancel waveforms in different channels and there is calibration that uses gain and offset matching between channels:Measuring biological signals is very challenging. You are trying to address many different obstacles.
Power supply noise is the least of your problems. You are surrounded by 60 Hz AC line interference.
You need very high CMRR, common mode rejection ratio.
Are you using instrumentation amplifiers?
View attachment 353489
Hi Secan,I got only low price used but still I can't take chances to build 7.2V and bring down to 5V unless it is commercially available). I
Is there a commercially available 7.2V to 5V clean regulator? Ill just buy one to try. Dont want to experiment building it bec it would take $4000 to repair the $17700 unit (see description in last message).Hi Secan,
I am only suggesting you try the 7.2V 5V reg method in order to check IF the noise you are having problems with, is due to the Boost switcher.
It may not be the source of the problem.
E
I think the TS's reply in post #8 (" When it is battery powered. There were no vertical lines.") is proof that it is use of the AC adapter which is the source of the problem.I am only suggesting you try the 7.2V 5V reg method in order to check IF the noise you are having problems with, is due to the Boost switcher.
Are you saying to achieve 3uV resolution. You should be able to acquire a current of V/R = 0.000001/100,000,000 = 0.03 picoA in the signal? Is 0.03picoA possible? What corresponding signal impedance is it?Ohm's Law.
If your input current is 1 nA, you have a signal of 100 mV.

I've been pondering the above this whole sunday and realised what you meant.Ohm's Law.
If your input current is 1 nA, you have a signal of 100 mV.
What will power the active electrode? You would need a power source which itself does not introduce further interference.If one has active electrode
Yes and no.I've been pondering the above this whole sunday and realised what you meant.
1. We don't measure skin current in EEG but voltage
2. So the 1nA you mentioned is not from the skin (I thought from the skin) but from the EMI interference inducing current in the electrode wire. Hence you need lowest source impedance. Ok got this if this is what you meant.
3. If one has active electrode which has electrode output impedance of just 10ohm to avoid attracting huge EMI interference. What is the use of electromagnetically shield room or faraday cage when you can just use active electrodes? Because I read some experiments done in faraday cages.

My g.USBamp has no RIGHT LEG (DRL) electrode. Instead you need to put the ground near yours ears with the bipolar pair in the head. I have the g.GAMMAbox active driver and active electrodes. The electrode has very high input impedance and only 10ohm output impedance so it attracts little interference in the wire. Also the g.USBamp doesn't use instrumental amplifier but 16 ADCs where they substract the common mode between ADCs (channels) by firmware or software. There is calibration by means of gain and offset mismatch equalizer between channels that takes care of any slight impedance etc mismatch between the ADCs.Yes and no.
Biological electrical sources have very high source impedances. If the input impedance of the measuring instrument is low, you end up with a reduced voltage signal. You want the input impedance to be high to avoid the voltage divider effect.
The common mode signal is the unwanted interference picked up by the long leads of the EEG cables. This is eliminated or reduced by using an instrumentation amplifier with high CMRR (common mode rejection ratio).
Noise from the power supply is eliminated or reduced by choosing op amps with high PSRR (power supply rejection ratio).
Since the EEG signals into the inputs are virtually floating, you need to reference them to a 0 V point. This is the purpose of the RIGHT LEG electrode.
View attachment 353541
Reference: https://openeeg.sourceforge.net/doc/modeeg/modeeg_design.html
(My masters thesis measured and analyzed EEG signals by applying FFT.)
But if the differential amplifiers or ADCs have CMRR that is very high. All the 60Hz interference can be cancelled, isn't it?Your biggest enemy is 60 Hz interference coming from AC electrical supply. That is why you need a Faraday cage and operate off batteries.