Please help fixing this thing! PEMF machine

thatoneguy

Joined Feb 19, 2009
6,359
Ok, I'm confused then.

The resistor, 820, what is the voltage across it? This allows finding current through it.

The 3V on D8 is a problem, as Zeners aren't made that small.

If the speaker has output, and there are pulses on the other lines of the NAND gate, that should be switching the transistor on and off as well, unless the pulses never overlap.
 

thatoneguy

Joined Feb 19, 2009
6,359
let me check....what about the resistor in between D7 and D8?
I missed it in the zoom, rewrote post.

Now looking for voltage across that resistor, which should be about 9V, to get an idea of current.

If you do a quick temporary short of CE on the 2N2219, does the output pulse once?
 

Thread Starter

petit

Joined Jan 24, 2013
70
Now looking for voltage across that resistor, which should be about 9V, to get an idea of current.
yes, 9.30V

The 3V on D8 is a problem, as Zeners aren't made that small.
yeah, i think there is a problem around that area, but i get stuck there drawing, there is a op amp that i cant find the datasheet

If you do a quick temporary short of CE on the 2N2219, does the output pulse once?
when i did that, the speaker beeped, like adding and "extra pulse" but i couldnt check for pulses on the pads. let me check that
 

thatoneguy

Joined Feb 19, 2009
6,359
That simply checks to ensure the transformer functions, it will be a short output pulse, so put the meter in peak (min/max) mode if it has one.

Hopefully the meter is fast enough to catch it.

If there an output, the pulse transformer is good, and the problem lies upstream of it, if there is no output, then the transformer isn't working, but that doesn't mean something else isn't wrong, since the transistor isn't getting turned on.
 

Thread Starter

petit

Joined Jan 24, 2013
70
If there an output, the pulse transformer is good, and the problem lies upstream of it, if there is no output, then the transformer isn't working, but that doesn't mean something else isn't wrong, since the transistor isn't getting turned on.
i dont understand why is not working shorting E-C, even if the meter is not fast enough to catch the pulse, at least the Red Led should turn on with that pulse
 

thatoneguy

Joined Feb 19, 2009
6,359
i dont understand why is not working shorting E-C, even if the meter is not fast enough to catch the pulse, at least the Red Led should turn on with that pulse
Which one is the red one? LED1 should be on all the time when power is applied.

LED2 also relies on some other components being good, transformer, SCR, inductors, etc.
 

Thread Starter

petit

Joined Jan 24, 2013
70
Which one is the red one? LED1 should be on all the time when power is applied.

LED2 also relies on some other components being good, transformer, SCR, inductors, etc.
LED2 is the one that pulses with the pads, if the pads get signal, that led should blink, it shares output
 

thatoneguy

Joined Feb 19, 2009
6,359
Then the pulse transformer or something after that point isn't functioning.

That's why I asked to see if there was any output on the secondary, it's simply following the signal to the output and seeing where it stops.
 

Thread Starter

petit

Joined Jan 24, 2013
70
Then the pulse transformer or something after that point isn't functioning.

That's why I asked to see if there was any output on the secondary, it's simply following the signal to the output and seeing where it stops.
but the readings on that pulse transformer looks good...should i go ahead and replace that transformer? if i can find it a replacement...:confused:
 

thatoneguy

Joined Feb 19, 2009
6,359
Is there a pulse on the secondary when a voltage is applied on the primary?

I don't suppose you have a signal generator?

Circuits like this tend to require a signal generator and scope to get to the problem quickly, e.g. to test the output stage, remove transistor, add signal generator, see if output is good. If it isn't, most of the circuit is eliminated as a major fault. If it is, then a scope is used to follow the signal.

Trying to do this with somebody else with only a multimeter is... difficult, but possible. I don't know how it functions or why, but I can show you basic troubleshooting skills.
 

thatoneguy

Joined Feb 19, 2009
6,359
I was stating what would happen in ideal circumstances.

Right now, it's just a matter of finding out if the transformer shows a pulse on the secondary or not, if not, ensure there is a voltage across the primary when C-E is shorted if there isn't one there, it's backing up a bit.
 

Thread Starter

petit

Joined Jan 24, 2013
70
Right now, it's just a matter of finding out if the transformer shows a pulse on the secondary or not, if not, ensure there is a voltage across the primary when C-E is shorted if there isn't one there, it's backing up a bit.
i dont understand exactly how that pulse transformer works. pin1 has 16V and is also connected to colector, pin2 also 16V, pin3 is ground and pin4 is output??
 

thatoneguy

Joined Feb 19, 2009
6,359
The pulse transformer works the same way a regular transformer works.

The difference is pulse transformers are designed to be 50% or less duty cycle, and more friendly to square wave input. Standard transformers "prefer" sine wave input due to the function of inductors. Pulse transformers are constructed the same way, but with intent to lower effects of inductance so a square wave pulse (on/off) can pass mostly undistorted.

A pulse on the primary results in a pulse on the secondary of a different magnitude (depending on the turns ratio).
 