I bought a handheld o-scope. It says Maximum input: 10X: 600 Working Voltage. Does that means I can hook it up to 240V line voltage to see if I am getting dirty electricity?
Yes and no.I bought a handheld o-scope. It says Maximum input: 10X: 600 Working Voltage. Does that means I can hook it up to 240V line voltage to see if I am getting dirty electricity?
PLUS .. if you have grid tied solar panels, or any neigbour (every 3rd), then that can add sheetloads of noise !Regarding seeing the voltage as "dirty": as mentioned before, the outlet rarely has a pure and clean sinewave as motors, switching power supplies, CFL lightbulbs and other highly non-linear loads introduce a number of interferences (harmonics) to the lines.
I wasn't aware of the uncleanliness of solar panel inverters. Thanks!PLUS .. if you have grid tied solar panels, or any neigbour (every 3rd), then that can add sheetloads of noise !![]()
It is the Jinhan JDS2022A. The input resistance is 1Mohm/10Mohm +/-2%. Does that say anything?As per reply, it can get VERY dangerous !! The DSO "Earth" will be either Neutral or LIVE !! depending on house wiring. You can use an isolation
transformer, even better, with say10:1 reduction - or make up a PURE resistance divider, again with 10:1 ratio.
If you want to check current, there are plenty of pick up coils for that. There are other tricks as well.
What is the make / model of the DSO?
Oh, OK. Well, it is no big deal to wire up a 40VA transformer and step it down from 240V to 24V. I have fried many of them. But I don't get why I can probe 240V with my multimeter but can't with the oscilloscope.Oscilloscope's input impedance is usually 1MΩ. This changes to 10MΩ with a 10X attenuation probe. Your scope max input is 40V, or 400V with 10X probe.
It would be advisable to connect the probes via an isolated step-down transformer, for example, 240VAC to 12VAC transformer. This will give you the isolation protection as well as reduced voltages
It is a matter of safety.Oh, OK. Well, it is no big deal to wire up a 40VA transformer and step it down from 240V to 24V. I have fried many of them. But I don't get why I can probe 240V with my multimeter but can't with the oscilloscope.