Drok and others sell digital 200 micro amp meters but I have been unable to find any reference to the input impedance of such devices. Obviously, low is good, high is bad--anyone had an experience with one?
Thanks.
Thanks.
What's your idea of high and low?Obviously, low is good, high is bad
It's not so simple as low = good, high = bad. It's a tradeoff.Drok and others sell digital 200 micro amp meters but I have been unable to find any reference to the input impedance of such devices. Obviously, low is good, high is bad--anyone had an experience with one?
Thanks.
Forgive my asking but why would you want a meter for measuring voltage with a 1.0 K Ohm input impedance? We started with a current meter with a range of 0 to 200 uA which is a 0.2 mA meter and really a pretty common current meter. Now based on the above we went to a 200 mV meter which while also common I can't understand a 1,000 Ohm input impedance as that would tend to load the source as a volt meter. I assume you found a 200 mA meter with a 1,000 Ohm input impedance and not a 200 uA or 200 mV meter? Even the latter makes no sense as the voltage drop across the input impedance (200 mA * 1,000 Ohms)?Well, after a long time googling I found a meter 200 mV meter which was listed as having a 1 k resistance.
Thanks for all your help.
Yes, and this is what it comes down to as WBahn points out. In the above case the down side is I have a burden voltage drop across my current measuring plane of 200 mV or I loose .2 Volt that I can't get back. Enter the trade off. What if I am measuring the current in a 3.3 Volt circuit? Can I afford a 0.200 Volt drop on my current measuring plane? So I have to compromise somewhere to find that sweet spot.Digital meters generally require 200mv for full scale readout. So a 200uA readout would require a 1000 ohm shunt,correct?
SG
It's not so simple as low = good, high = bad. It's a trade off.
Most current meters are designed so that the voltage drop across them is about the same on each scale. This means that the resistance goes up as the range goes down. For 200 uA, expect a pretty high resistance. It's a tradeoff. As the resistance goes down, the voltage goes down and the signal-to-noise ratio goes down. Higher resistance, better measurement, up until the point that the resistance starts affecting the measurement.
So there's a sweet spot and what that sweet spot is depends on the application -- how accurate/precise do you need the measurement against how much perturbation can you accept in the system while making the measurement.