MisterBill2
- Joined Jan 23, 2018
- 27,888
REturning to the original complaint in post #1,which appears to be power frequency harmonics in the output signal, and tha complaint being that they can not be magicaly extracted somehow. At this point my conclusion about the problem is a bit different, and likewise my suggestion for a solution.The fiirst suggestion that I offer is to power that entire expensive amplifier package from a battery supply that is not connected to the mains in any way during the amplifier use. My second suggestion is to use a connection to the patient scheme similar to the scheme usedfor connecting heart monitors for use during surgery. THAT SCHEME uses two attachments to a differential input AND a third nearby attachment for the common reference voltage for that input pair. AND to use for tthose input connections fairly short unshielded wires with the majority of the connection being in a single, shielded cable with the reference common to the shield at one point only. THAT approach will reduce the pickup from the rather high noise voltage field in the patient vicinity.