Hypatia's Protege
- Joined Mar 1, 2015
- 3,228
Yes, please do!Indeed No effect in the stated position 'EXT X". I measured with the probe! I guess I safely can repeat with the DMM directly in order to have more exact values?
Yes, please do!Indeed No effect in the stated position 'EXT X". I measured with the probe! I guess I safely can repeat with the DMM directly in order to have more exact values?
With DMM directly: 103V; no effect at all of R69Indeed No effect in the stated position 'EXT X". I measured with the probe! I guess I safely can repeat with the DMM directly in order to have more exact values?
Are you certain you are measuring between pins #10 and #12 -- as opposed to #3 and #1???With DMM directly: 103V; no effect at all of R69
I ment: 156/C5 = 156/C55=C55= 4.23->4.12V (varies with R69);all ref to frame:
At 156/C5 (after TR31)
C55: 4.23->4.12V (varies with R69);
F: 108: (no effect of R69)
G: 4,8V: (no effect of R69)
pin 10 to chassis
CCW:0V
CW:0V
pin 12 to chassis:100;
no change with R69
Fantastic !!CU tomorrowVery interesting!!! -- Now we're getting somewhere!
So here's the deal -- I must be going for today - In the meantime I'll study the schematic in view of the test results and get back to you tomorrow!
Very best regards
HP![]()
Are you absolutely certain that you're measuring pins #10 and #12???pin 10 to chassis
CCW:0V
CW:0V
pin 12 to chassis:100;
no change with R69
--Please check for collector to emitter conduction also (CE should test ∞ in both directions) While, in theory, this should be unnecessary -- practice, however, doesn't always agreeI checked TR 33 and TR37 again BE and BC 0.7; infinite in reverse.
indeed it is possible that I d'nt disconnected from the mains... but on the other hand the scope is now in the state in which it was before the soldering. So there must be another cause.Was the scope connected to the mains when you unsoldered/resoldered R68 and R69? If so, it is possible that the electrically 'leaky' soldering iron 'altered' a semiconductor component --- Again, I'll have a close at the schematic and get back to you latter on--- Those deflection plate readings are very suggestive/interesting!
Best regards
HP![]()
Yes they are;Do you have ready access to test points 156/G and 156/F ??? (see uper right of schematic in post #114)
If so please verify that 156/G is connected to CRT Socket Pin #10 and 156/F to CRT Socket Pin #12
I concur with your assesment that the Scope's return to 'partial life' pretty much rules out such damage!indeed it is possible that I d'nt disconnected from the mains... but on the other hand the scope is now in the state in which it was before the soldering. So there must be another cause.
This is truly bizarre! -- Ok... Now that the scope is "working" again -- please check check X1 and X2 (pins #10 and #12) to each other and to chassis ground again --- Please remember to set the 'Time/Div' control to 'Ext X' prior to the testYes they are;