Hi,
I have a 100Hz 1V (peak-to-peak) square wave on a function generator, I have my true RMS multimeter plugged into it and I meausure around 0.5 V_RMS. So far so good!
Now I change it’s duty-cycle to 10% and I measure around 0.3 V_RMS.
Does this make sense?
My question is, if my multimeter reading is correct, how would I get to that value by the definition of RMS?