I have a GDS1052U oscilloscope from GW Instek. In the manual is mentioned about the probe compensation.
Does it matters if I make the probe compensation with the coupling on AC or DC ?
Does the wave forms from the attached screenshots look good ?
 

Attachments

crutschow

Joined Mar 14, 2008
34,469
It may not make any difference for a 1kHz square-wave, but using the DC coupling is best.

Those screenshot waveforms both look good.
 

Thread Starter

mike69

Joined Dec 22, 2012
83
Thanks.
So it should be good if I made the compensation using AC coupling ?

PS I already made some important measurements and I need to know if the probes were correctly compensated using AC coupling.
 

Reloadron

Joined Jan 15, 2015
7,523
Yes.

Here are a few examples of an old analog scope.

Normal Compensation:
Normal Comp.png

Over Compensation or High Frequency Accentuation:
Over Comp.png

Under compensation or High Frequency Attenuation:
Under Comp.png

Ron
 

Danko

Joined Nov 22, 2017
1,835
Does it matters if I make the probe compensation with the coupling on AC or DC ?
Theoretically you should make probe compensation with coupling on DC only.
It is important to have in view that with AC coupling signal always is distorted:

Calibrating.png
 
Last edited:

Thread Starter

mike69

Joined Dec 22, 2012
83
I re calibrated the probes using DC coupling.
Does the screenshots look good ?
I used 2 different probes.

I alreade made, yesterday, some important measurements (power supply noise and ripple and oscillation), so I need to know if the probes were correctly compensated.
 

Attachments

Thread Starter

mike69

Joined Dec 22, 2012
83
I did not changed any compensation settings from the probe.
More exactly, I did not rotated any screw on the probe.
 

Danko

Joined Nov 22, 2017
1,835
Theoretically you should make probe compensation with coupling on DC only.
I alreade made, yesterday, some important measurements (power supply noise and ripple and oscillation), so I need to know if the probes were correctly compensated.
My post was not only about compensation, but mostly about measurements with AC coupling.
With AC coupling measurements, measured signal give off its direct component, and obtains distorted form on low frequencies.
For power supply noise and ripple and oscillation measurement use AC coupling is just essential.
 