How to build a Voltage limiter for a millivolt signal?
anoopak
Joined Nov 24, 2023
77
I had built a millivolt conditioning circuit. A 0-40mV differential DC signal is fed to AD620 with a set gain of 100. The output of AD620 [0-4V] is fed to ADS1115 . Is there any method to limit the millivolt input to AD620 below 50mV so that ADS1115 will get saved when a higher input is accidentally fed to the input of AD620? Millivolt signal is generated from a Wheatstone bridge circuit [Common mode voltage ~10V].
ericgibbs
Joined Jan 29, 2010
21,622
hi,
You could consider reducing the supplies to say +/-5V.
The AD620 output saturates at approx Vs-1.2V, so the ADC could never be overdriven.
tonyStewart
Joined May 8, 2012
239
the AD620 limit is ; Differential Input Voltage 25 V. No protection needed.
the ADS111x limit is ; Analog input voltage AIN0, AIN1, AIN2, AIN3 GND – 0.3 VDD + 0.3 V
The input protection diodes are current limited
Thus using series R between AD620 (+/-15V ) to ADS111x (+5,0V) you need to limit the current by 15V-5V = 10V or 10V/10mA using >= 1k resistors or 2k with margin. This is easier than say using 2.5V +/-2.5V on the AD620 but is also possible then the AD620 is referenced to 2.5V instead of 0V.
This requires your Wheatstone bridge to be +ve and unipolar so that the AD620 output is between the limits for the ADC.
if your bridge is bipolar
, then the AD620 needs to be biased with a positive output, preferably to 2.5V using REF in with say 10k to +5 and 0V buffered by any Op Amp. (gain=+1) to drive REF.
anoopak
Joined Nov 24, 2023
77
hi,
EView attachment 313525
I selected the value of power supply voltage based on Diamond plot tool [Analog devices] recommendations. It says that for a common mode voltage of 10V , power supply voltage of +15v and -15V is required for AD620 for its operation.
anoopak
Joined Nov 24, 2023
77
I operate ADS1115 with a gain of one where the maximum input voltage is 4V. I fear if any spike at the AD620 input can expose ADS1115 to voltages more than 4V which can damage ADC [ADS1115 supply voltage is 5V].
anoopak
Joined Nov 24, 2023
77
This means I don't have to worry about damage of AD620 [Bridge excitation voltage is 24V] .
Connecting a resistor of 2K between AD620 and ADS1115 is convenient option for me. Whether there will be any drop in signal voltage due to this resistor?
ericgibbs
Joined Jan 29, 2010
21,622
V1 is not a voltage source but some unknown values of a Wheatstone bridge, but may work if they are low enough R values. We need to see the bridge datasheet link. Also STP cables ought to be used to reduce stray line noise, preferably with PE gnd connection to shield.
hi tony,
V1 is a simulation test Voltage source that sweeps from 0mV through 10mV.
Knowing the bridge data is not required for this type of simulation.
E
ericgibbs
Joined Jan 29, 2010
21,622
You recommends connecting a 4V Zener diode at AD620 output? Is R3 is the current limiting resistor for Zener? Can you tell the purpose of R1 & R2? [Sorry I am not from Electronics background]
Hi anoo,
R1 and R2 are there to bias the AD620 inputs to Vsupply/2, which is typical for a bridge input
So with a +15V bridge supply that's 15v/2 =7.5v.
R3 is a current limit resistor to the zener diode, when the Vout of the AD620 rises above ~4v.
E
tonyStewart
Joined May 8, 2012
239
No voltage drop with next to no input current in datasheet.