Fixing an RF generator.

Thread Starter

mbohuntr

Joined Apr 6, 2009
446
Afternoon all,
I am refurbishing an EICO 330 RF generator for my bench, I ordered it from EBAY. I had the expected dirty rotary switches, and stuck pots. What I found was I think this was a kit, or somebody has "Repaired" it.

The board is fine, caps all appear good, no leaks or bulges, and after I cleaned the switches, and freed up the pot, I got all 5 bands producing a signal. the generator has 5 ranges from kHz to 54 Mhz with a 400 Hz modulated wave built in the circuitry

My issue is the 16MHz -54MHz band. The varicap and vatiable inductor have been gutted, and it has an Air coil with no cap at all??? At 16MHz, there is virtually no output, it peaks at around 30MHz, and puts out way under the 300millivolts it is supposed to on high. At 54 MHz, it is around half of the peak of 30MHz. somewhere around 20mV P-P.

If I wish to replace the inductor, and Varicap, I do not know the value of the inductor. The cap is 1-8pF, but no value for the inductor. I'm guessing around .2uH, but don't know for sure and what Q should I go for, low or high? Should I assume the resonance to 30MHz, and with a low Q get a wide band that attenuated at either and?

The schematic is enclosed, somebody changed it..The L5 inductor is now a hand wound coil, tapped at 2 turns, and the varicaps do not follow the schematic, the one that goes to L5 is gutted, no innards, and disconnected. The other outputs are strong and accurate to my Hantec scope, I put a 10 Meg pulldown resistor in the C band across the inductor to get it oscillating, it works much better now. I found similar slugged inductors, on line, I just do not know the proper value. Any ideas?

http://www.nostalgickitscentral.com/eico/schematics/eico_schematic_330.pdf
 
Last edited:

sghioto

Joined Dec 31, 2017
8,634
Are there any values marked on the other inductors? Can you see the windings and tell how many turns there are? If so you may be able to calculate the value based on that.
SG
 

Thread Starter

mbohuntr

Joined Apr 6, 2009
446
No values, only a part number that does not come up on a search...
Are there any values marked on the other inductors? Can you see the windings and tell how many turns there are? If so you may be able to calculate the value based on that.
SG
Eico cat . Part number 36085...
 

Thread Starter

mbohuntr

Joined Apr 6, 2009
446
Hello,

Here is the complete manual, wich includes some calibration tips.

Bertus
Thanks! It says to tune the inductor and cap to zero beat, but someone gutted both. There is no cap connected other than the main tuner, and the inductor has no core, and looks like a 16 ga. air core..
I was thinking that if I took the mid point of the range... say 35 MHz, and found the resonant frequency capacitance at the mid point of the tuner. (295pF) so 150 pF, I should be able to reverse engineer the inductor value.. Or am I completely off the mark?
SG... Heathkit SG6 has a similar coil set up, and frequency range value, but no values.... *surprised*
 

Thread Starter

mbohuntr

Joined Apr 6, 2009
446
Thread bump... I tried various combinations in a colpitts oscillator on my bench, the best I was able to come up with is 2) 68pf caps and an 1uF inductor giving me around 6-7 Mhz... When I put a 300pf air tuning cap in place of the lower cap, you only get around 20Khz of adjustment before it drops out. I have an acute lack of knowledge here..... I cannot get anywhere near the 16-54 Mhz adjustment. needed..

Update, after doing some more circuit tracing, It appears the band uses C9, and C3 in parallel to adjust the frequency, and the inductance range is .18uH - .29uH I reverse calculated it using a resonant frequency calculator.. Does anyone know if this is correct? Right now it uses a hand made air coil, and no 1-8pF cap.
 