Conducted EMC lab earthing?

Thread Starter

eem2am

Joined Jul 22, 2012
40
Hi,
We have a 2m by 2m room totally clad in alu sheet for our Conducted EMC testing of offline SMPS <100W.

Would it be a good idea for us to drive an earth rod into the floor of the lab and earth the alu sheets to that?...or would it be better to just earth the alu sheets to the earth wire in the lab's mains supply that supplies the Spec An and LISN etc?
 

Thread Starter

eem2am

Joined Jul 22, 2012
40
Aluminium was cheaper? The Earth port of the LISN must be hard-shorted to the earth plane, as such, the metal sheeting should be earthed?
 

Thread Starter

eem2am

Joined Jul 22, 2012
40
Thanks, though i actually mostly believe that the ground rod wouldnt make much difference......it is just important that the alu sheeting is earthed to the earth wire in the supply that supplies the LISN. Though if that earth wire has already snaked its way say 100m through a building full of electrically noisy things, and then come to the conducted lab, then i wonder how that puts it?
 

tautech

Joined Oct 8, 2019
496
Thanks, though i actually mostly believe that the ground rod wouldnt make much difference......it is just important that the alu sheeting is earthed to the earth wire in the supply that supplies the LISN. Though if that earth wire has already snaked its way say 100m through a building full of electrically noisy things, and then come to the conducted lab, then i wonder how that puts it?
Substandard IMO.
Better to setup a dedicated Earth/ground.
 