Thanks for the input, I previously tried replacing Q850 with a new 2n3904 transistor, and the old one tested fine after I had it out. So.. If Q850 is indeed the problem then it must be something about how it's getting fed.Problem with LM317T.
40 volt limit or the chip blows up.
I'm not too "proper" to redesign the 30 volt power supply, I just needed to point out the voltage limit on the chip.
I believe a 60 volt version is available.
CR841 will not "rectify" in the manner you suggested, however, if it is shorted, that would be a way to explain the whole problem. Lift a leg and check.
ESR is not a likely candidate for this problem. The capacitor would have to look like it's open before this kind of fault would cause this much error. (Don't worry about not having an ESR meter.)
Referring to post#4, I can see why you can't get zero volts on the base of Q850. As long as the +30 and -30 aren't balanced, your adjustment range will not center. The problem is that this adjustment should cause the balancing to happen, and it isn't. Q850 is responsible for pulling the down base of Q840 and it isn't. This makes Q850 a prime suspect. Either it's open or it isn't being fed properly.
LM317HV
I've checked most of the components that I think would have an impact on the current flow through Q850 and they check "OK" with my relatively primitive methods.