Hello everyone,
I'm working on measuring AC current using a current transformer (CT), the CT I’m using has a transformation ratio of 50:1, with a maximum current rating (Imax) of 50 A.
I’ve experimented with several burden resistors (4.7Ω, 1.2Ω, and 0.22Ω) and connected a logic analyzer to visualize the waveform and obtain the Vrms values.
Schematic:

Waveform (using a 4.7Ω burden resistor):

The waveform corresponds to an expected primary current of approximately 6.21 A. However, when calculating the primary current, I get:
Iprim = Isec *50 = (0.378/4.7) * 50 = 4 A
also tested another CT, the SCT-013, which has a 30A/1V ratio and comes with a built-in burden resistor. After repeating the same steps, I obtained this waveform:

So Iprim = Vrms * 30 = (212,806/1000) * 30 = 6,38
This result is much closer to the expected value.
Could anyone help me understand the discrepancy in measurements with the first CT? Any guidance would be greatly appreciated!
Best regards
I'm working on measuring AC current using a current transformer (CT), the CT I’m using has a transformation ratio of 50:1, with a maximum current rating (Imax) of 50 A.
I’ve experimented with several burden resistors (4.7Ω, 1.2Ω, and 0.22Ω) and connected a logic analyzer to visualize the waveform and obtain the Vrms values.
Schematic:

Waveform (using a 4.7Ω burden resistor):

The waveform corresponds to an expected primary current of approximately 6.21 A. However, when calculating the primary current, I get:
Iprim = Isec *50 = (0.378/4.7) * 50 = 4 A
also tested another CT, the SCT-013, which has a 30A/1V ratio and comes with a built-in burden resistor. After repeating the same steps, I obtained this waveform:

So Iprim = Vrms * 30 = (212,806/1000) * 30 = 6,38
This result is much closer to the expected value.
Could anyone help me understand the discrepancy in measurements with the first CT? Any guidance would be greatly appreciated!
Best regards