The 3 core settings for successful FFT are the timebase initially to have many periods on the display, the Start/Center/Stop frequencies and the FFT points for speed of FFT refresh.So where is the sample rate and how do you set it?
Knowing the fundamental frequency helps us make revealing settings so to display several harmonics we need set a center frequency of 2 or more times the fundamental.
Initially playing with FFT of the 1kHz probe compensation output lets us get the feel of making good FFT settings which we can apply to other real frequencies of interest however always the fundamental guides us how to make good settings.
Quickly, we already know the compensation output is 1kHz and to display this at the 1st graticule set a center frequency of 5kHz (3rd harmonic) and a Hz/div setting (horizontal scaling) of 1kHz/div. Square waves harmonics rise every odd fundamental multiple whereas sinewave harmonics rise every even fundamental multiple.
1000X-E series Siglents all offer 1Mpts FFT and to process all those can take some time and we don't always need such detailed FFT so we can reduce FFT points lower to speed processing and FFT sweep.
Often 128kpts is plenty to get adequate results. The FFT box on the display tells us how many are selected and in the case of the images you posted there are 1048576 = 1 Mpts.