Op-Amp Offset Voltage
yeezhihao
Joined Sep 30, 2007
11
Hi
Can someone explain to me why in this circuit, I would get an Output offset voltage in magnitude of volts?
I heard it was something along the lines that since input is 0v the equivalent circuit is that both capacitors can be removed together with the feedback line. And since 22m is huge, no current will flow to ground bla bla bla..........but I just cannot put everything together. Can someone please be kind enough to explain to me??
Thanks in advance!
Mark
ecb123
Joined Mar 6, 2008
7
check the specs on your op amp-bias current, op-amps need certain requirments
-22M too big-your circuit acts like a simple diff amp with these values.read up on op-amps
yeezhihao
Joined Sep 30, 2007
11
Hi ECB123
Would you care to provide me with more details?
Thanks a lot!
Rgds
Mark
rwmoekoe
Joined Mar 1, 2007
172
inputs have bias current (pretty small though). this current requires to be passed through equal resistor values for each inputs, to cancel offset at the differential inputs.
in your case, one input gets the 22megs while the other gets only 22k//27k or 12k! imagine the different in voltage at the inputs, times the openloop gain of the opamp... no wonder the volts at the output are there.
Audioguru
Joined Dec 20, 2007
11,248
You show the resistors as 22m ohms which is 0.022 ohms which is almost a dead short.
I think you mean 22M which is 22 million ohms.
The lousy old 741 opamp has too much input bias current to use resistors with such a high value. You need an opamp with FET inputs. They have no input bias current, just a little leakage current.
nomurphy
Joined Aug 8, 2005
567
You're making a high-pass filter with Fc = 0.33 Hz, is that really necessary?
But, at least lower the resistance values and raise the cap values by a factor of 100. That would be 220K and 2.2uF.
With a FET op amp you might get away with a factor of 10, that would be 2.2Meg and 0.22uF.
(I prefer 22Meg or 0.022, over 22M or 22m, to help avoid confusion.)
yeezhihao
Joined Sep 30, 2007
11
Hi rwmoekoe
Thanks for the explanation.
I get what you mean totally.
Thank u for your help.
yeezhihao
Joined Sep 30, 2007
11
Hi Audioguru
Yes, I meant 22Mega. Sorry for the confusion.
Thanks for your recommendation.
Rgds
yeezhihao
Joined Sep 30, 2007
11
Hi nomurphy
Yes, its a high pass filter with cutoff at 0.036.
As I'm building a ECG aquisition device I would like to make the signal as clean as possible, thats why I have this HPF even though such low frequency noise many not be common.
I have increased my cap to micro values.
Rgds
rwmoekoe
Joined Mar 1, 2007
172
yeezhihao,
you don't have to change the values of the existing components.
just add a 22meg res between the inverting input and the 27k // 22k node. the input
will 'see' the impedance as 22meg (plus 12k :) ) also.
or, calculate a new combination to replace the 27k // 22k that results in a 22meg or near as parallel. (56meg // 47meg = 26meg). well, it's better the first one though, 'cept for the component count, if that matters