Can i measure 450nA with a 6000 count DMM or am i better off just doing a voltage measurement across a resistor and employing ohms law?
I have tried to use a DMM but can't get it to respond.
panic mode
Joined Oct 10, 2011
5,308
that is 0.5uA so not really... proper thing would be to build suitable amplifier. but... you can try using volts range. most DMMs have input impedance of 10MOhm on most ranges. so 0.45uA*10MOhm = 4.5V. so switch DMM to 20V range (0.45uA will display as 4.5V).
if 4.5Vdrop is too much, place 1.1MOhm resistance (1Meg + 100k in series) in parallel with DMM. that will result in voltmeter input resistance drop from 10 to 1 MOhm so voltage would be 10x lower. for 0.45uA you would be seeing 0.45V
something like
schmitt trigger
Joined Jul 12, 2010
2,260
Be aware though, that many autoranging DMMs change their input impedance between 10 and 11 Megohm, as they step through their different ranges. Check the meter’s user manual for range/impedance info and lock it into a manual range.
But if 10% error is close enough for your requirements, then you don’t need to bother.
Homebrew1964
Joined Nov 22, 2024
289
I am using a PM18 DMM (pictured) which is fully manual at the 60uA range and i get nothing, i tried to measure 100mA on another range and still nothing, i think the fuse is blown for the current ranges and i can't get the thing open so i'll either trash the meter or just use it for the other functions. Another meter i tried works fine.
WBahn
Joined Mar 31, 2012
33,282
Have you looked in the battery compartment? That is typically where the fuses are located on handheld meters.
Homebrew1964
Joined Nov 22, 2024
289
The instructions are incorrect, the fuses are inside the unit on the circuit board, i had to take the outer Orange sleeve off the meter and undo another 4 screws to get to them, the 600mA fuse was blown and i replaced it with a 400mA one, put it all back together and it works as new so thanks for your replies people.