In addition to Dana's fine advice and employing the use of an instrumentation amplifier I see your voltage divider being used to provide the excitation voltage to your load cell. Since the output is a function of the excitation voltage (mV/V) you want to make absolutely sure that your excitation voltage is stable. Less the IA you may want to measure the load cell output with a known load and observe the output over a period of time to make sure it is stable and remains stable. Another forum member has a quite lengthy thread on the subject including maintaining a good zero with the amplifier over time. You can read the thread here.