New to the forum, and I'm going to start this post by saying I'm a mechanical engineer with no formal electrical training. Try to not laugh too hard if the problem is obvious
I had a circuit set up to test a unit at work. The basic circuit was as follows:
120 VAC 2 prong power cable ---> Bridge rectifier ~
Bridge rectifier + ---> Multimeter to measure current ---> Large coil ---> Bridge rectifier -
Plugging this into the wall worked fine. The coil received rectified DC voltage and current was about what I expected.
Then, I tried to set up an oscilloscope to look at the voltage waveform the coil was receiving. I was told to use an isolating transformer on the oscilloscope for this purpose to prevent a ground loop. I plugged the 'scope into the isolating transformer, connected the probe to the + coil lead and the ground clip to the - coil lead. As soon as I plugged in the 120 VAC power cable connected to the rectifier, there were sparks and a breaker tripped. There were no visible shorts, as in alligators touching or anything like that.
Later, I tried plugging the rectifier power cable into the isolating transformer and plugging the scope into the non-isolated wall socket, and this time, it worked. What changed?
I read some other posts on similar topics. One in particular caught my eye, which was a post by MrChips in the thread "At what point Isolation transformer needed":

New to the forum, and I'm going to start this post by saying I'm a mechanical engineer with no formal electrical training. Try to not laugh too hard if the problem is obvious
I had a circuit set up to test a unit at work. The basic circuit was as follows:
120 VAC 2 prong power cable ---> Bridge rectifier ~
Bridge rectifier + ---> Multimeter to measure current ---> Large coil ---> Bridge rectifier -
Plugging this into the wall worked fine. The coil received rectified DC voltage and current was about what I expected.
Then, I tried to set up an oscilloscope to look at the voltage waveform the coil was receiving. I was told to use an isolating transformer on the oscilloscope for this purpose to prevent a ground loop. I plugged the 'scope into the isolating transformer, connected the probe to the + coil lead and the ground clip to the - coil lead. As soon as I plugged in the 120 VAC power cable connected to the rectifier, there were sparks and a breaker tripped. There were no visible shorts, as in alligators touching or anything like that.
Later, I tried plugging the rectifier power cable into the isolating transformer and plugging the scope into the non-isolated wall socket, and this time, it worked. What changed?
I read some other posts on similar topics. One in particular caught my eye, which was a post by MrChips in the thread "At what point Isolation transformer needed":
To my understanding, I was doing the initial option to "float" the scope, but I still tripped a breaker. The second option is the only one that ended up working. Any insight into why this happened would be greatly appreciated. I would like to understand what caused it for future reference. Thank you in advance for your patience with this mechanical engineer!Here is another example of where you could use an isolation transformer.
Most if not all oscilloscopes are grounded and hence voltage measurements are taken with reference to GND.
Now suppose you wanted to measure the voltage across two terminals using the probe and the ground clip. That would not be a good idea because you would be grounding your circuit through the ground clip.
You use an isolation transformer to take the scope off ground, i.e. your scope is now "floating". Now you can use the ground clip as your reference since it is now a floating reference.
Alternatively, you could power your test circuit (power supply and everything connected to it) via the isolation transformer and achieve the same effect.