The 330 Ohm resistor serves to limit inrush current at the moment when the mains is applied. Diode bridge can withstand 50 Amp surge, so you can easily minimize the resistor. If a resistor of 47 Ohm is used, power dissipations is P=I^2*R= 0.05A^2*47=0.12W only. Current surge is 240V/47= 5 Amp only, so everything is OK with the bridge.That calculation makes sense, (even though I still don't understand the concept of impedance )
That solves the first of my initial questions.
I presume that R12 is ignored due to it's 1M value
The second one was about,, 'when a fault develops', the circuit is supposed energize both relays. I put a 20K resistor in R11, but It still wont switch.
With all respect to 'Ian0' and I believe it was 'Hymie' they suggest that the Current Transformer is the wrong type and they suggest a different one eg ZMCT103C , has a hole of 5mm. But.....I can't run a 240 volt lead through a 5mm hole.
I wonder what your opinion is?
I'll try that tomorrow the current CT ..To buy an other CT takes 4 weeks to come from ChinaThe more primary (i.e. mains side) turns you can get on the core, the better it will work. Get a CT with a large hole in the middle and fill it.
Wont the impedance be vanishingly small at 50/60 Hz?What is the secondary inductance?
That appears in parallel with the load resistance, so the voltage across the load resistors (6.8k in your case) is the current multiplied by the parallel combination of the inductor reactance and the load resistor.
That's one reason why a CT is normally operated with a very small load resistor (the "burden"), so that the current through the inductance is negligible compared with the current through the load.
If that were the case, it wouldn’t work. The majority of the output current would go through the internal inductance.Wont the impedance be vanishingly small at 50/60 Hz?