That's more like a victim resistor...What a arcing 100,000V 8kW power supply does to a damping resistor.
View attachment 246531View attachment 246532View attachment 246533View attachment 246534
People buy flukes because they tend to be repeatable and consistent.Excellent containment of arc damage and fire suppression. This is why companies buy expensive Fluke meters.
Do you mean a Fluke or a fluke look-alike? I've never seen an unfused 10A input on a fluke. Please let us know which model fluke he was using.10amp un fused plug
I'm pretty sure that any fluke that does have a 10A input has a fuse. A relatively expensive fuse on my Fluke 85 and 87 - like $12 each.My Fluke 27 has a 10A MAX input. But it IS fused. All inputs are fused.
Sure, arc flash originates at the meter probes. No meter can stop that from happening. Repeatable and consistent over long periods of time is right on the money but meter safety is a large consideration to limit meter originated dange when screwups happen.People buy flukes because they tend to be repeatable and consistent.
A test or trouble shooting procedure written 25 years ago for some legacy equipment is a real pain in the arse if you can't get a fluke 77 anymore to do the tests.
Companies pay for a fluke because they are slow to change on the surface so the old test procedures remain the same and inside the electronics package from one to another is very reliable as the units age.
Also they don't blow up in your face, but....
A good friend of mine was doing some live tests on a GE silpac drive.
It was a real monster running a 7000 hp motor at 700 vdc.
My chum made the mistake of no looking at the meter and seeing it was on the 10amp un fused plug ( reasons never explained ).
He touched something live on the valve stack and the arc flash happened.
His face and hands were pretty badly burned.
We had some arc flash boundary of approach calculations done.
We went from working on it live to not being allowed near it when it was running unless you had a 40 cal bomb suit on or were more than 20 feet from it.
So its not just the meter its what you are working around.
I do like you board repairs.
The other input is high impedence (10M Ohms) for voltage so there is no need for a fuse at rated voltage levels. Other higher-end meters have fourth terminal for milliAmp/microamp inputs and a 400mA to 600mA fuse. There are, essentially no infused inputs on the 115 meter.My Fluke 115 is also marked 10A / FUSED
Since this instrument only has the 10 Amp full scale current measurement range, there appears to be no fuse on the other Red Socket.