I wanted to know what does a multimeter measure if it is a sine wave or any other waveform which is not a DC waveform?
I suggest you give this a read as it addresses your question and points out Crest Factor as mentioned by danadak. Yes, some DMMs use micro controllers but as I mentioned, all of this depends on the meter and manufacturer. The better the meter, the more features, the higher the cost. This is also true of hand held as well as bench models. When choosing a meter you choose what you want based on your specific criteria. Here is an example using an inexpensive meter average responding RMS indicating and a True RMS responding RMS indicating meter.What kind of internal circuit they use for this? For example they need to measure the time difference like sample time t1 and sample time t2 etc. I feel it is very complicated. Do they use any internal micro controller to do the processing or normal discrete components?
But it is not obvious which of the two meters is reading the correct rms voltage of the modified sine wave.I suggest you give this a read as it addresses your question and points out Crest Factor as mentioned by danadak. Yes, some DMMs use micro controllers but as I mentioned, all of this depends on the meter and manufacturer. The better the meter, the more features, the higher the cost. This is also true of hand held as well as bench models. When choosing a meter you choose what you want based on your specific criteria. Here is an example using an inexpensive meter average responding RMS indicating and a True RMS responding RMS indicating meter.
Measuring a clean RMS sine wave, US mains power:
View attachment 152063
The same two meters measuring a MSW (Modified Sine Wave)
View attachment 152064
The meter on the left is an inexpensive about $12 USD meter, the meter on the right is about a $400 USD meter. The difference is pretty obvious.
Ron
The reference to a "modified sine wave" implies that the output is likely from a UPS or inverter and is attempting to output a sine-like waveform at about 120 Vrms.But it is not obvious which of the two meters is reading the correct rms voltage of the modified sine wave.
Depends on the meter. Remember, AC measurements with handheld meters long predate digital multimeters.What kind of internal circuit they use for this? For example they need to measure the time difference like sample time t1 and sample time t2 etc. I feel it is very complicated. Do they use any internal micro controller to do the processing or normal discrete components?
That’s being very presumptuous – I would be interested to see what reading the Fluke meter gives when fed with a half wave rectified (mains) sine wave.The reference to a "modified sine wave" implies that the output is likely from a UPS or inverter and is attempting to output a sine-like waveform at about 120 Vrms.
I'd be willing to take that bet.That’s being very presumptuous – I would be interested to see what reading the Fluke meter gives when fed with a half wave rectified (mains) sine wave.
I’d be willing to bet it won’t read close to 85V (based on 120Vac mains).
Because of previous measurements I’ve made (of half wave rectified mains) using various meters (including a Fluke Trms meter).I'd be willing to take that bet.
The Fluke 87 is a well-respected True-RMS responding meter and a halfwave rectified signal at mains frequency is well within its capabilities.
Why do you believe it is otherwise.
Time to buy a new meter?My Fluke 77 reads 68v AC with a simple lamp load half wave rectified.
Max.
I should add that that was about the same response I got – getting twice that value based on 240Vac.Time to buy a new meter?
I am putting my money on the more expensive true RMS Fluke 87 pictured on the right in my image.But it is not obvious which of the two meters is reading the correct rms voltage of the modified sine wave.
Measuring an everyday 9 V battery on the AC ranges the meter on the right reads Zero V and the meter on the left reads something like 19 V which is what we should expect.Fluke believes that RMS meters should display only the AC component (may be exceptions! - read the manual), so they "AC-couple" the input to the meter - there is a capacitor that blocks any DC component. A Fluke meter set to AC will read zero volts for a pure DC input. To be sure of what the actual RMS value is, you must measure on the DC range, then on the AC range, and calculate the actual RMS value. You do this by squaring each component, adding the squares and taking the square root of the sum. The RMS value of DC is equal to the average of the DC.
If what ebp says is correct (and I’ve no reason to doubt it), any ac waveform with a dc offset will result in a multimeter giving an erroneous Trms value.The previous posts point out something very important: you must know how any instrument you are using works.
Note - every numeric value in the following is approximate.
A pure, undistorted sine wave of 120 volts RMS has a peak voltage of 168 volts. If you half-wave rectify the sine, you get something that has both an AC component and a DC component. The DC component is 53.5 volts. The AC RMS component is 64.8 volts.
Fluke believes that RMS meters should display only the AC component (may be exceptions! - read the manual), so they "AC-couple" the input to the meter - there is a capacitor that blocks any DC component. A Fluke meter set to AC will read zero volts for a pure DC input. To be sure of what the actual RMS value is, you must measure on the DC range, then on the AC range, and calculate the actual RMS value. You do this by squaring each component, adding the squares and taking the square root of the sum. The RMS value of DC is equal to the average of the DC.
So
(53.5^2 + 64.8^2)^0.5 = (2862 + 4199)^0.5 = 84 volts RMS
Sometimes it is nice to have a blocking capacitor in the meter, sometimes it is an annoyance. It depends on what you are doing. But you must know how your meter works!
Max didn't report the peak or RMS for his unrectified voltage, but his meter is likely working just fine.
At first I was shaking my head at this, but IIRC (and I had forgotten this since it has been so many years since I've used a Fluke for AC measurements of non-sinusoids and even that was for a very specific short-lived purpose) the Fluke meters report the RMS value of the AC component only. I don't know their rationale for doing so, but if you want the actual RMS value of the waveform you need to take the Pythagorean sum of the DC and AC measurements.My Fluke 77 reads 68v AC with a simple lamp load half wave rectified.
Max.