Hello! I have a UNIT-T UT210E clamp meter, which I think it has a problem regarding the current measurements.
I will describe the problem below: when I measure current, if the wire is positioned into the center of the clamp (the center marked with the 2 arrows) then it measure 10.16 Aac. But, when I move the wire inside the clamp, then the measured value modifies like this: in the upper part of the clamp the value is 10.5 Aac and in the lower part of the clamp (the closest part to the body of the clamp meter) the value is 9.95 Aac.
I made a calibration of the clamp meter, but the results didn't change.
Meanwhile, if I measure the current using a standard multimeter set to 20Aac then the measured value is 9.9 Aac.
I would like to know if the deviation generated by moving the wire inside the clamp is normal or not ?
LE: in the manual it shows the accuracy on the 20A range as +-(2.5%+8), I understand the +-2.5% but I don't understand how to read the "+8".
I will describe the problem below: when I measure current, if the wire is positioned into the center of the clamp (the center marked with the 2 arrows) then it measure 10.16 Aac. But, when I move the wire inside the clamp, then the measured value modifies like this: in the upper part of the clamp the value is 10.5 Aac and in the lower part of the clamp (the closest part to the body of the clamp meter) the value is 9.95 Aac.
I made a calibration of the clamp meter, but the results didn't change.
Meanwhile, if I measure the current using a standard multimeter set to 20Aac then the measured value is 9.9 Aac.
I would like to know if the deviation generated by moving the wire inside the clamp is normal or not ?
LE: in the manual it shows the accuracy on the 20A range as +-(2.5%+8), I understand the +-2.5% but I don't understand how to read the "+8".