Toggle switches and IR/DS testing

Thread Starter

harpazo

Joined Sep 20, 2017
8
Hello friends. I am designing a test fixture that will be used for dielectric strength and insulation resistance testing. It will be used for multiple units used on commercial aircraft. My plan is to use 35 toggle switches that will be used to select which wires are in circuit for DS and IR testing. The voltages typically used for these tests will be 1.2KVAC and 500VDC. I've been looking at the S303 SPDT switch from NKK Switches but I'm uncertain as to how their switch voltage ratings will apply in my application. The switches are rated for 15A@125VAC, 6A@250VAC, and 20A@30VDC. As you know, the current in view when doing DS and IR testing is very small, no higher than 5mA.

Will the application of 1.2KVAC and 500VDC from a Hi-Pot tester destroy these switches?

Thanks for your input.

Mike
 

bertus

Joined Apr 5, 2008
22,278
Hello,

What are IR and DS testing?
There is a great chance of arcing when you use a switch on a higher voltage than rated.

Bertus
 

Thread Starter

harpazo

Joined Sep 20, 2017
8
IR is insulation resistance testing. DS is dielectric strength testing. Arcing is one of my concerns and that's why I am looking for answers as to what those voltage ratings on the switches actually mean. I was trying to extrapolate the numbers using Ohm's law and am actually still toying with that but I want to make sure I have the right information on the switches before I spend the money for all the switches and build it only to find out it's not going to work. I hate when that happens. :)
 

eetech00

Joined Jun 8, 2013
3,961
Hello friends. I am designing a test fixture that will be used for dielectric strength and insulation resistance testing. It will be used for multiple units used on commercial aircraft. My plan is to use 35 toggle switches that will be used to select which wires are in circuit for DS and IR testing. The voltages typically used for these tests will be 1.2KVAC and 500VDC. I've been looking at the S303 SPDT switch from NKK Switches but I'm uncertain as to how their switch voltage ratings will apply in my application. The switches are rated for 15A@125VAC, 6A@250VAC, and 20A@30VDC. As you know, the current in view when doing DS and IR testing is very small, no higher than 5mA.

Will the application of 1.2KVAC and 500VDC from a Hi-Pot tester destroy these switches?

Thanks for your input.

Mike
You'll have to look at the insulation breakdown characteristics of each component used to route the Hi-Pot test voltage to the DUT(including any wires). The switches should have a rating specified on their data sheet.
Beyond the datasheets, the only really way to know is to test the fixture with the selected components.

eT
 
You cannot apply 1200 VAC to a 250 VAC switch and expect a good outcome!

The high voltage relays have the contacts sealed in an inert gas. This will extinguish the arc that occurs when the contacts open. Using a switch in air will ionize the air and may have a sustained arc. It can also burn the plastic insulation and cause a conductive track between the contacts.

Always put safety first and using under rated switches is not safe!
 