[SOLVED] Tektronix 2213 Oscilloscope

Thread Starter

Homberg

Joined Apr 11, 2025
13
I have a vintage Tektronix Oscilloscope that I purchased in 1982. I was trying to use it on a project that I was working on, and found that although I was able to calibrate a 10X probe on channel 2, no adjustment of the probe (that worked on channel 2) really changed the appearance on the scope to look like a square wave.

I tried many local and non-local electronics repair operations, and found none that could fix it for me. In the mean time, I had purchased a "new" oscilloscope that was faster and relatively cheap.

I would still like, for mainly sentimental reasons, to get my old Tektronix in working order. I have attached a portion of a schematic from the service manual.

If anyone is willing to help me, I can provide other schematic jpegs and videos of the oscilloscope screen (I couldn't attach the videos to this post).
 

Attachments

t_glover

Joined Mar 16, 2021
56
I had a couple old Iwatsu oscilloscopes that had similar issues. A small amount of Deoxit in all the switches did wonders for both scopes.
 

MrChips

Joined Oct 2, 2009
34,700
It is not clear to me what is the problem.
You have a Tektronix 2213 oscilloscope. It works on CHAN1 and CHAN2.
You have a 10X probe.
Has the probe been properly adjusted on CHAN1 and CHAN2?
 

Futurist

Joined Apr 8, 2025
732
I have a vintage Tektronix Oscilloscope that I purchased in 1982. I was trying to use it on a project that I was working on, and found that although I was able to calibrate a 10X probe on channel 2, no adjustment of the probe (that worked on channel 2) really changed the appearance on the scope to look like a square wave.

I tried many local and non-local electronics repair operations, and found none that could fix it for me. In the mean time, I had purchased a "new" oscilloscope that was faster and relatively cheap.

I would still like, for mainly sentimental reasons, to get my old Tektronix in working order. I have attached a portion of a schematic from the service manual.

If anyone is willing to help me, I can provide other schematic jpegs and videos of the oscilloscope screen (I couldn't attach the videos to this post).
A picture's worth a thousand words...
 

Thread Starter

Homberg

Joined Apr 11, 2025
13
I have a spray can of DeoxIT that I've attached. Is this the stuff you are talking about?

I have seen more comments that don't think that using DeoxIT will work. It's a simple enough thing to try, so I will.

I'm wondering how I can calibrate the probe to show a nice looking square wave on channel 2, but have no adjustment on channel 1. Did you just spray every switch?
 

Attachments

MrChips

Joined Oct 2, 2009
34,700
What does the trace look like with 10X probe on CHAN 1?
What does the trace look like with 10X probe on CHAN 2?

The adjustment is in the 10X probe. Have you adjusted this?
 

Thread Starter

Homberg

Joined Apr 11, 2025
13
As I said the trace on channel l 2 was calibrated to show a nice square wave The same probe on channel 1 would not calibrate. I could not attach video files to this forum, so I can only say that it would not calibrate to anything close to a square wave. I am willing to email the video file to anyone that wants it.
 

MrChips

Joined Oct 2, 2009
34,700
As I said the trace on channel l 2 was calibrated to show a nice square wave The same probe on channel 1 would not calibrate. I could not attach video files to this forum, so I can only say that it would not calibrate to anything close to a square wave. I am willing to email the video file to anyone that wants it.
This is what I understand.
10X probe on CHAN2 shows square wave.
10X probe on CHAN1 does not show square wave.

Post photos of oscilloscope screen and scope settings for each case.
Don't send video files. Take photos of the screen and post photos.
 

Reloadron

Joined Jan 15, 2015
7,870
What are you tweaking? The Probe or the vertical input channel of the scope?

Probe compensation looks like this:

IMG_5849.JPGIMG_5850.JPGIMG_5851.JPG

Top High Frequency Accentuation.

Middle Critical Comp (Just Right)

Bottom High Frequency Attenuation.

What are you using for the input signal?

Ron
 

Thread Starter

Homberg

Joined Apr 11, 2025
13
I tried spraying the DeoxIT that I had into the 2 black (dark) portions of the attenuator switch assembly behind channel 1, and it worked!!! (See Photo Attached)

I originally took videos of the oscilloscope screen, because I had the screen showing both traces, so I couldn't take a picture of the screen When I displayed the trace for channel 1 or channel 2 only, I was able to take screen shots.

After spraying, I get a square wave on both channels (See Attached Photos)! Unfortunately, I don't have a before picture of the problem. Anyway, that doesn't matter now.

Thanks to everyone who responded to my postings, especially t_glover!
 