I started to work with the following oscilloscope: Multicomp pro Digital Storage Oscilloscope 100MHz MP720665
I used a function generator: PM 5127 PHILIPS to generate functions. (Yes, it is pretty old.)
I executed the self calibration and also the factory reset.
I gave a 2000 Hz sinusoidal wave and the result is around 1366 Hz.
How can I solve this issue?
Thanks a lot.
35,245
You cannot perform a proper calibration on either instruments without a known frequency standard.
A simple solution is to get a quartz crystal oscillator module that generates an accurate frequency.
ericgibbs
Joined Jan 29, 2010
21,601
Hi Slan,
I would say looking at the very poor form factor of the Sine wave, it's a low grade sine wave generator.
I would suspect the calibration of the generator scale.
E
Ian0
Joined Aug 7, 2020
13,297
Measure the output of a mains transformer. What frequency does it read? and what frequency should it be?
Danko
Joined Nov 22, 2017
2,241
You can easily check digital oscilloscope by measure
very accurate period (2 sec) of pulses from ordinary wall clock:
Slania87
Joined Jul 9, 2025
9
Hi!
Thanks a lot.
Slania87
Joined Jul 9, 2025
9
Hi!
Thanks a lot.
Slania87
Joined Jul 9, 2025
9
Thank You for the answers. I attached the measured square wave as well.
Unfortunately, my hands are tied, I cannot buy another oscilloscope.
35,245
Hi!
Thanks a lot.
What is the problem?
35,245
Of course I can see that. So which unit is in error, the oscilloscope or the signal generator or both?
...waiting for TS to respond.
tautech
Joined Oct 8, 2019
507
... that the 2000 Hz signal is processed as 1366 Hz by the oscilloscope.. ?
Not at all surprising from one of these:
For more precise frequencies you require a digital AWG that can provide Sine, Square and Pulse that you can adjust to be a defacto Mark Space generator.
Back in the days when I used to fix scopes I had a PM5131 FG and it was a horrible unstable and inaccurate bit of kit totally unsuitable for checking/adjusting scopes.
A SDG805 would be vastly superior but frequencies will never perfectly match unless both the AWG and scope can be externally referenced.
bertus
Joined Apr 5, 2008
23,013
The test in #15 will give an 50 or 60 Hz signal on the scope dependent on your countries mains voltage frequency.
Bertus
28,454
Frequency measurement with an oscilloscope, if it is done by measuring the time between zero crossings, is often subject to setup and operation errors. What is the scheme that your scope uses???