I've used the LT1793 in numerous designs requiring low input bias current and high input resistance, but it requires at least a ±5 volt supply to operate properly; you won't be able to use it on a single +5V supply. I would strongly suggest using one of the op amps mentioned in the document @Danko linked to.oohhhhh gotcha gotcha ok. I think my next steps are to aquire a correct op amp and see where i go from there
I think so. If you already have one on hand I would give it a try.Found a TLV2471
I have been and no results so far. I'll keep at it thoughI think so. If you already have one on hand I would give it a try.
What do you mean exactly by no results?I have been and no results so far. I'll keep at it though
Agreed, to monitor pH you will need a hiZ multimeter. Some handhelds have a HiZ fumction. Manufacturers like Keyence anf Keithley make them multimeters with high impedance inputs."The source impedance of a pH electrode is very high because the thin glass bulb has a large resistance that is typically in the range of 10 MΩ to 1000 MΩ.
This means that the electrode can only be monitored by a high-impedance measuring device"
See solution here
View attachment 218018
Buy this cheapest, but powerful PH meter and connect your probe to its circuit.Hi all, I am trying to determine whether my pH probe is giving me accurate data or not. I was hoping to measure the voltage with a multimeter. I have the probe in a pH buffer of 4 and according to pH to voltage scale. 1 pH is about 59 mV/pH, Starting at 7 pH = 0. IF thats the case i should be seeing around 150mV but i only get about 10mV