Hi all,
I've measured a phase current of BLDC using Fluke i310s current probe and oscilloscope, but now I'm having trouble converting the current probe output to an actual current through the phase. Before the measurement, the probe was adjusted to zero and set to the 300 A range with 1 mV/A. The probe attenuation set in the scope is 10, the vertical scale is 0.1 V. How do I convert the value of 0.6 measured on the oscilloscope channel to the current?
Somehow my interpretation of the current isn't what I expect it to be, therefore I'd like to verify this with you.
Thank you.
I've measured a phase current of BLDC using Fluke i310s current probe and oscilloscope, but now I'm having trouble converting the current probe output to an actual current through the phase. Before the measurement, the probe was adjusted to zero and set to the 300 A range with 1 mV/A. The probe attenuation set in the scope is 10, the vertical scale is 0.1 V. How do I convert the value of 0.6 measured on the oscilloscope channel to the current?