I bought a second hand oscilloscope a couple of weeks ago. It had an issue, the one mentioned in the topic, but at least it did display something so I thought I would probably be able to fix it. It now became somewhat of a challenge for me and I spent a lot of time with it and learned tons about that old art of analog circuits. Amazing engineering!
I did some research and came to the conclusion that the error would probably come from the later y-amplifiers because it would affect both channels in the same way. I opened it and noticed that someone has already worked on it, and not very diligently. One of the bridge rectifiers had been replaced by an ugly huge one which didn't fit at all aka "3D-legsbending". I also noticed some modifications on the Z-board: some resistors have been added that did not only have a different value than given in the datasheet, but also were connected in a completely different logic. Also, an additional capacitor was added somewhere on the back side of the PCB.
First research attempts:
After some research on this specific model, I checked the checkpoints and all except one were okay (ChP 7, see in the picture below). I also measured the voltages (those less than 2kV) at the CRT and they were fine except for the y-deflection plates (~55V but should be 85) More general information on how to troubleshoot an oscilloscope lead me to the idea to short the bases of the final y-amp's transistors (yellow line in picture). I hope I managed to isolate the fault to the second stage of transistors within that circuit (red area in picture). Lo' and behold, it was the same one where all that tinkering had been done!

However, while working on it the error magically disappeared so I suspected a loose connection, which made perfect sense considering the looks/dust of them and how hard it was to get them apart and together. The display was also indicating that the knobs and potentiometers were very noisy so I decided to do a complete restoration of my "lucky cheap repair".
What I did:
- Replaced all electrolytic capacitors
- Restored original parts wherever I saw some hacks (green resistors in picture)
- Cleaned the heck (!) out of it. Yes, it took tens of hours. However, I did not touch the trimmers.
Result:
Okay, so now I got a nice shiny oscilloscope with smooth potentiometers and so on, but the error came back :'( I still notice somewhat of an improvement, though in terms of signal distortion. So the problem that I can only use the upper half of my scope remains. I can use y-Pos and put my signal above the center to work around it but it is still kind of distorted/nonlinear. When I put 2 identical signals on top of each other they almost look the same but when they have different y-position they will also have a different y-size. Everything that is lower than 0 will be displayed as 0 so I can use the calibrator signal and make it a flat centered line by simply reducing the y-position enough.Further investigation:
So now that I know where the tinkering had been done I became a lot more suspicious about the Z-Board. Especially since a checkpoint voltage was wrong (~5.2V but should be 3,75V) coming from between the y-final amp rails. Only then I found out that the faulty section does not even contain the original transistors anymore but some other ones with different specs (red in picture). No wonder it's behaving strangely...
The picture shows the display in dual mode. Both probes are connected to the calibrator but one channel's y-position is turned down. The line is my "absolute zero" and it's all that is left of the square wave.

Future plans / call for support:
The next step is that I wanna replace the transistors with the original models. But I somehow got the idea that it is not the only thing going wrong in here... that's why I am asking for your help. Soeone with more experience can probably help me out here. I will try to provide you with all the information I can get so just ask if you have an idea.
What I don't wanna do:
I don't want to measure anything in the kV range (but then again I don't think that section is where the problem is located anyway). Even more important, I don't want to buy a cheap digital one instead because "it's not worth repairing" or "they are better anyway". I know that and I don't care about it. I already paid more for the probes and parts, shipping, DeOxit and so on. I am not poor so I could just buy a new one but I am doing this for fun and to learn more about electronics. I don't expect to have a 100% accurate Hm 203-4 in the end that works until 20MHz. All I wanna measure with it are audio signals so 20kHz in a quantitative fashion so it's fine for me to sacrifice some accuracy for nostalgia.
Any help would be greatly appreciated =)
Introduction:
I bought a second hand oscilloscope a couple of weeks ago. It had an issue, the one mentioned in the topic, but at least it did display something so I thought I would probably be able to fix it. It now became somewhat of a challenge for me and I spent a lot of time with it and learned tons about that old art of analog circuits. Amazing engineering!
I did some research and came to the conclusion that the error would probably come from the later y-amplifiers because it would affect both channels in the same way. I opened it and noticed that someone has already worked on it, and not very diligently. One of the bridge rectifiers had been replaced by an ugly huge one which didn't fit at all aka "3D-legsbending". I also noticed some modifications on the Z-board: some resistors have been added that did not only have a different value than given in the datasheet, but also were connected in a completely different logic. Also, an additional capacitor was added somewhere on the back side of the PCB.
First research attempts:
After some research on this specific model, I checked the checkpoints and all except one were okay (ChP 7, see in the picture below). I also measured the voltages (those less than 2kV) at the CRT and they were fine except for the y-deflection plates (~55V but should be 85) More general information on how to troubleshoot an oscilloscope lead me to the idea to short the bases of the final y-amp's transistors (yellow line in picture). I hope I managed to isolate the fault to the second stage of transistors within that circuit (red area in picture). Lo' and behold, it was the same one where all that tinkering had been done!

However, while working on it the error magically disappeared so I suspected a loose connection, which made perfect sense considering the looks/dust of them and how hard it was to get them apart and together. The display was also indicating that the knobs and potentiometers were very noisy so I decided to do a complete restoration of my "lucky cheap repair".
What I did:
- Replaced all electrolytic capacitors
- Restored original parts wherever I saw some hacks (green resistors in picture)
- Cleaned the heck (!) out of it. Yes, it took tens of hours. However, I did not touch the trimmers.
Result:
Okay, so now I got a nice shiny oscilloscope with smooth potentiometers and so on, but the error came back :'( I still notice somewhat of an improvement, though in terms of signal distortion. So the problem that I can only use the upper half of my scope remains. I can use y-Pos and put my signal above the center to work around it but it is still kind of distorted/nonlinear. When I put 2 identical signals on top of each other they almost look the same but when they have different y-position they will also have a different y-size. Everything that is lower than 0 will be displayed as 0 so I can use the calibrator signal and make it a flat centered line by simply reducing the y-position enough.Further investigation:
So now that I know where the tinkering had been done I became a lot more suspicious about the Z-Board. Especially since a checkpoint voltage was wrong (~5.2V but should be 3,75V) coming from between the y-final amp rails. Only then I found out that the faulty section does not even contain the original transistors anymore but some other ones with different specs (red in picture). No wonder it's behaving strangely...
The picture shows the display in dual mode. Both probes are connected to the calibrator but one channel's y-position is turned down. The line is my "absolute zero" and it's all that is left of the square wave.

Future plans / call for support:
The next step is that I wanna replace the transistors with the original models. But I somehow got the idea that it is not the only thing going wrong in here... that's why I am asking for your help. Soeone with more experience can probably help me out here. I will try to provide you with all the information I can get so just ask if you have an idea.
What I don't wanna do:
I don't want to measure anything in the kV range (but then again I don't think that section is where the problem is located anyway). Even more important, I don't want to buy a cheap digital one instead because "it's not worth repairing" or "they are better anyway". I know that and I don't care about it. I already paid more for the probes and parts, shipping, DeOxit and so on. I am not poor so I could just buy a new one but I am doing this for fun and to learn more about electronics. I don't expect to have a 100% accurate Hm 203-4 in the end that works until 20MHz. All I wanna measure with it are audio signals so 20kHz in a quantitative fashion so it's fine for me to sacrifice some accuracy for nostalgia.
Any help would be greatly appreciated =)