I am trying to troubleshoot Fluke 8842A DMM. If anyone has knowledge of this particular unit please share your thoughts.
No blown fuses both front and back and inside the unit.
AC power supply to VFD ( Vacuum Florescent Display) reads 4.1 Volt at Transformer pins 14, 16. ( 4.5 v nominal )
Power supply to Display IC voltages all have power (DC Power supply voltages )
VFD physically not damaged.