I'm attempting to calibrate a precision current source, @ 1A, and 3A. I have a Fluke 87v, and a Brymen 857.
I also have a precision 100 mOhm resister, 0.1% tol., 10ppm, 3W:
https://www.digikey.com/en/products/detail/vpg-foil-resistors/Y14750R10000B9R/8538226
When I adjust my current output for say, ~3A, and measure across the resistor, I'm reading 300mV. Because it's 0.1% shunt, my slop might be 300.3mV, or 299.7mV. Wattage under test is 0.9W, well within this shunt resistor's specification.
Now, the % difference between the ideal 300mV, and a possible outlier of say 300.3 is = 0.1%, as expected. However, my Fluke in current measurement reads 2.978A, which is .73333333% difference, when it should read no more than 0.2% deviation, according to it's specifications. My Brymen 857 current reading is 2.9761A, .79666666%, again way outside of the meters specifications.
This is the current spec. for the Fluke 87v:
Scale: 6A, Resolution: 0.001A, Burden: 0.03 V/A, Accuracy: ±(0.2 % + 4)
I don't know who to trust! I paid nearly $30 for that resistor, is it lying to me?
I also have a precision 100 mOhm resister, 0.1% tol., 10ppm, 3W:
https://www.digikey.com/en/products/detail/vpg-foil-resistors/Y14750R10000B9R/8538226
When I adjust my current output for say, ~3A, and measure across the resistor, I'm reading 300mV. Because it's 0.1% shunt, my slop might be 300.3mV, or 299.7mV. Wattage under test is 0.9W, well within this shunt resistor's specification.
Now, the % difference between the ideal 300mV, and a possible outlier of say 300.3 is = 0.1%, as expected. However, my Fluke in current measurement reads 2.978A, which is .73333333% difference, when it should read no more than 0.2% deviation, according to it's specifications. My Brymen 857 current reading is 2.9761A, .79666666%, again way outside of the meters specifications.
This is the current spec. for the Fluke 87v:
Scale: 6A, Resolution: 0.001A, Burden: 0.03 V/A, Accuracy: ±(0.2 % + 4)
I don't know who to trust! I paid nearly $30 for that resistor, is it lying to me?