Hi,
I have what should be a simple circuit but it is not behaving as expected.
A power supply is connected directly to a load ( used for a MAS spectrometer). The power supply is 900V DC with a user variable current of 30-200nA. I don't know anything about the load ( so black box ). The manufacture won't tell me where in the power supply circuit the current is being measured but we suspect it is not right at the output. I would like to measure the current coming from the power supply to the unit.
I put a resistor in series between the power supply and the unit (on the low side). Using a battery powered DMM I wanted to measure the voltage across the resistor to get the current. The DMM has a resolution of 0.01mV on the 100mV range ( Keysight U1241C).
From Ohm's Law I expected that 50nA through a 1k would give: ( V=50*10^-9 * 1*10^3 = 0.05mV). I also tried 10k ( V=0.5mV ). For both I get a reading of 0.
Should I just try a 100k ( or higher) or is there something else I'm missing? Is there something in the DMM data sheet I missed? As I increase the value of the resistor, I will have to start taking into account the resistance of the meter to be able to make accurate current readings.
For mV AC the sensitivity is 15mV ( assuming it has to be higher than 15mV before it shows a reading) but there is no mention for DC.
Would there be a better way to measure the current?
Thank you!
I have what should be a simple circuit but it is not behaving as expected.
A power supply is connected directly to a load ( used for a MAS spectrometer). The power supply is 900V DC with a user variable current of 30-200nA. I don't know anything about the load ( so black box ). The manufacture won't tell me where in the power supply circuit the current is being measured but we suspect it is not right at the output. I would like to measure the current coming from the power supply to the unit.
I put a resistor in series between the power supply and the unit (on the low side). Using a battery powered DMM I wanted to measure the voltage across the resistor to get the current. The DMM has a resolution of 0.01mV on the 100mV range ( Keysight U1241C).
From Ohm's Law I expected that 50nA through a 1k would give: ( V=50*10^-9 * 1*10^3 = 0.05mV). I also tried 10k ( V=0.5mV ). For both I get a reading of 0.
Should I just try a 100k ( or higher) or is there something else I'm missing? Is there something in the DMM data sheet I missed? As I increase the value of the resistor, I will have to start taking into account the resistance of the meter to be able to make accurate current readings.
For mV AC the sensitivity is 15mV ( assuming it has to be higher than 15mV before it shows a reading) but there is no mention for DC.
Would there be a better way to measure the current?
Thank you!