Just a reminder THERE IS HIGH VOLTAGE INSIDE THAT BOX. Don't touch anything that looks rubbery or is covered with a lot of fine, dark dust. Please be careful and stay safe.As you can observe, CH1 goes between 0V and 75mV according to your settings.
CH2 is between -30mV and +30mV. Hence it is AC coupled.
I suspect that you have lost connection in the AC-GND-DC switching somewhere.
You will have to open the scope and start hunting around while using CH1 as a working scope.
We are here to guide you.
Where did you find the manual? I downloaded a copy buy it was scanned to a PDF and most of the schematics are cut off on the edges.With x10 probe and VOLTS/DIV set to 5mV we should be reading 50mV/div.
Hence I am seeing CH1 amplitude to be about 1.5 divisions = 75mV which is about x4 lower than 0.3V CAL OUT.
What does CH1 look like on DC vs AC? Does it look the same?
Download the SS-5702 Service Manual off the internet. We will be looking at the CH1 and CH2 input attenuation section for faults. CH2 is showing AC coupling (high-pass filter) with an RC time-constant of about 100μs, i.e. you have lost straight through DC coupling.
Set both CH1 and CH2 AC-GND-DC to GND and adjust the trace to some horizontal reference lines on the grid so that we know where is the 0V reference. Then set AC-GND-DC to DC and show us the actual DC levels with respect to GND.
View attachment 236916
Good eye. I did not notice that. Push CH2 POLARITY to NORM.Is channel 2 set to inverted ? I am asking this as channel 2 looks like a differentiated version of an inverted channel 1 waveform.
I agree with Yaakov in using channel 1 to trace the path through the channel 2 amplifier.
Les.
I would vote for investing some Deoxit contact cleaner (https://www.amazon.com/Hosa-D5S-6-DeoxIT-Contact-Cleaner/dp/B00006LVEU) which is very good and will help with the maintenance of good connections. Don't spray it all over the inside, just on the contacts of the wafer switches, then operate them several times. This is really good stiff in my experience. It's not just cleaner, it has a contact enhancer in it. It's like Stabilant ut much cheaper.In that case, ignore what I have posted and don't bother with taking more photos.
You have identified the problem with the CH2 AC-GND-DC switch.
Open the unit and inject or spray isopropyl alcohol on to the switch contacts and work the switch while the alcohol has not yet evaporated.
You should have your problem solved.
It’s safe but it purposely leaves a film behind so apply it only to the contacts and wipe it off other stuff if it gets there. The film isn’t harmful but it can attract dust and make a messI was going to ask about Deoxit and if it would be safe to use in this application.
I will see if I can get some today. Where I used to work before retirement
used to keep it in stock.