Differential probe displays approx. Twice actual voltage

Thread Starter

SiCEngineer

Joined May 22, 2019
439
Hello.
I am using the following Keysight probe: https://www.keysight.com/gb/en/product/N2792A/differential-probe-10-1-200-mhz.html to probe a gate drive output voltage. The expected output is +15V and -5V. However when checking into the oscilloscope with 1Meg and 10:1 attenuation, the actual signal shows as +30V and -10V. The probe has a 10x attenuation and is not configurable. It also has a +\~20V peak differential voltage. I have tested the power supply of the gate drive and it is as expected without switching enabled. I have not yet tested the voltages while the gate drive is actively switching.

is there any commonly known explanation as to why a probe may show 2x the voltage magnitude when connected to a scope? If so is there any way to eliminate this issue?
 
Last edited:

Thread Starter

SiCEngineer

Joined May 22, 2019
439
The 'scope you're using does have a 50Ω input impedance?
it does but I’m using the 1Meg input impedance. I checked this morning and the scope reads:

+36.8V
-11.6V

the multimeter is attached to the probe and reads:

+17.58V
-5.437V

the probe has an offset calibration mode which I attached to my 5V power supply but at minimum offset with the screwdriver it still reads that as about 8V. I tested again and the scope reads pretty much the same voltage as it did before the calibration.
 

Irving

Joined Jan 30, 2016
3,548
So I set the probe to 50R? I guess the input impedance of the probe is 1Meg from data sheet but it outputs 50R to the scope?
From probe datasheet:
" Both N2792A and N2793A probe are compatible with any oscilloscope with 50Ω BNC inputs"

Switch the scope to 50ohm, or use a 50ohm terminator between probe and scope.
 

Thread Starter

SiCEngineer

Joined May 22, 2019
439
From probe datasheet:
" Both N2792A and N2793A probe are compatible with any oscilloscope with 50Ω BNC inputs"

Switch the scope to 50ohm, or use a 50ohm terminator between probe and scope.
thank you Irving and Albert. I seem to have got confused with the input impedance of the probe and the scope.
 

MisterBill2

Joined Jan 23, 2018
16,582
It could be a setup issue such that the scope is set to report Peak to Peak while some other place it thinks it is reading Peak volts. Or Vice-Versus. Units display setup can lead to a lot of confusion if there are errors. So it might not be a hardware issue at all.
Usually hardware errors are just a small percentage off.
 

Thread Starter

SiCEngineer

Joined May 22, 2019
439
It could be a setup issue such that the scope is set to report Peak to Peak while some other place it thinks it is reading Peak volts. Or Vice-Versus. Units display setup can lead to a lot of confusion if there are errors. So it might not be a hardware issue at all.
Usually hardware errors are just a small percentage off.
Thanks Bill for this input - I did think it could be that as I’ve had same issue before. Fortunately setting the scope impedance to 50R resolved the issue though.
 