Hi Jim,This could not be solved by trimming right ?
That is not the way to trim.
Did you read the Page #1 & 2 section in the PDF regarding trimming>
Once you have trimmed the LC's in that way, it will not matter where within the area of the load bearing plate you place the load.
The combined Vout of the 4 LC's will be an averaged value of the 4 LC's individual measured weights.
An important point to consider is that if for example the individual LC was specified as 10mVout at say 100kg its maximum load weight, the averaged Vout for a 100kg weight on the 4 LC scale would be 10mV/4 = 2.5mV.
Give me details of the LC types and the expected max load weight.
What are youe weighing?
BTW: when you say: object will drop
I would advise not to drop weights onto the baseplate that could over stress the LC's.
Also, it is advisable to carry out a Calibration weight test after construction, and adjust the Voffset and Span of the weighing scale
E