ok, i now have to resort to asking for some help with this issue.
i previously posted asking for anyone that has a fluke 8010a or 8012a dmm to take some voltage readings for me, but no luck.
i have the service manual for this unit and the specs come from that.
below is the image of the power supply portion of the schematic.
the service manual calls for the voltages as printed in blue, but, these voltages are based on the non battery operated version. unfortunately, i have not been able to find any voltage specs for the battery operated version.
in 1984, an addendum was made that changed a resistor and a cap in the timing circuit and the addendum also made change that the table for voltages from the battery operated supply to meet the specs that are listed in blue on the image below, which is the same specs as the a.c. only meters.
the battery pack consists of 2 sub c cells for a total of 2.7vdc 2 ah.
voltage from the battery in on condition is 2.72vdc and delivers 150ma.
my concern is the + and - 15 volt spec.
i have scoped the outputs of q6 and q7 and duty cycle is approx 70% on. i will add pics of these later to help.
i am confused that fluke would call for the battery operated version to hold the same specs as the a.c. only version without having a prior table specifically for the battery operated version.
i think there is a problem somewhere as i played a little with r25 and noted that changing that resulted in the + and - 15 vdc coming into specs, BUT, upon turn off and turn on, the clock cycle would not start unless the r25 was within it's original spec of 4.3k.
can you guys give me a hand with this?
i'll update with pics of the scope later.
thanks
bill
i previously posted asking for anyone that has a fluke 8010a or 8012a dmm to take some voltage readings for me, but no luck.
i have the service manual for this unit and the specs come from that.
below is the image of the power supply portion of the schematic.
the service manual calls for the voltages as printed in blue, but, these voltages are based on the non battery operated version. unfortunately, i have not been able to find any voltage specs for the battery operated version.
in 1984, an addendum was made that changed a resistor and a cap in the timing circuit and the addendum also made change that the table for voltages from the battery operated supply to meet the specs that are listed in blue on the image below, which is the same specs as the a.c. only meters.
the battery pack consists of 2 sub c cells for a total of 2.7vdc 2 ah.
voltage from the battery in on condition is 2.72vdc and delivers 150ma.
my concern is the + and - 15 volt spec.
i have scoped the outputs of q6 and q7 and duty cycle is approx 70% on. i will add pics of these later to help.
i am confused that fluke would call for the battery operated version to hold the same specs as the a.c. only version without having a prior table specifically for the battery operated version.
i think there is a problem somewhere as i played a little with r25 and noted that changing that resulted in the + and - 15 vdc coming into specs, BUT, upon turn off and turn on, the clock cycle would not start unless the r25 was within it's original spec of 4.3k.
can you guys give me a hand with this?
i'll update with pics of the scope later.
thanks
bill