The Electrician
- Joined Oct 9, 2007
- 2,970
The result of getting a negative value is because using a series capacitor whose ESR is much larger than the battery impedance means that you are trying to get a small value (battery impedance of something like .0226 ohms) by subtracting one large number (capacitor Z) from another large number (battery + capacitor Z). This is a well known numerical problem that sometimes occurs when making physical measurements.So , as you said about the Z (of capacitor alone direct subtracting from the Z of the series )
if i get a negative value ? (i can take absolute one )
Not a problem??
This is why you will not get accurate results when using a series capacitor of only 1000 uF. If you can, use the 4700 uF capacitor; it should have a smaller ESR. And, of course, the very best solution is to avoid using a series capacitor at all--in other words, use the method with the external bias offset applied to the LCR meter and no capacitor.
If you cannot convince your tutor that the external bias method is better, be sure to note in your report that the accuracy limitations imposed by using a series capacitor are significant and unavoidable with the series capacitor method.
You should probably consult with your tutor as to the question about the introduction.