Scope offset while in AC coupling mode?
Elerion
Joined Sep 11, 2017
125
Hi everyone.
Is there any reason why an (almost new) oscilloscope in AC coupling mode shows a small offset (a couple of milli volts)?
Example (using 1x probe):
Is this normal for relatively cheap new oscillosopes?
I didn't realize this behaviour with anolog scopes.
The offset spoils the (AC) RMS measurements, as it takes into account a mean value, and it is annoying.
Thank you!
Deleted member 115935
Joined Dec 31, 1969
I dont know, I've not ussed a scope to measure RMS,
but
I'd have thought the RMS needs to know the DC offset of the signal,
so can you take an RMS measurement if the scope is AC coupled ?
Ian0
Joined Aug 7, 2020
9,815
It's because the frequency of the signal is too low to fit on the screen.
I'm guessing that you have about 500us per division so the screen is about 7ms wide, and that the signal you are measuring is perhaps 50Hz, so that the negative-going half is off the screen to the right.
dl324
Joined Mar 30, 2015
16,918
AC coupling mode shows a small offset (a couple of milli volts)? ...
What is the vertical sensitivity? DSO's are infamous for their fuzzy traces.
Analog scopes can also have an offset. Unlike DSO's, you don't need to go through a menu to remove the offset; you just turn the knob.
Just measure the peak value and divide by 1.414 to get RMS.
Ian0
Joined Aug 7, 2020
9,815
tautech
Joined Oct 8, 2019
386
Hi everyone.
Thank you!
Check the manufacturers websites for new firmware versions and install them as necessary.
After which and the scope has been running for some while so good and warm, run the Self Cal utility which should address AC coupling channel offset.
Better measurement accuracy can sometimes be got using measurement Statistics if your scope has such.
Ian0
Joined Aug 7, 2020
9,815
Just turn the timebase down to a slower speed, and the offset will be gone when you can see the rest of the waveform
Elerion
Joined Sep 11, 2017
125
Check the operating manual. There is usually a way to adjust the zero offset even in the most inexpensive scopes..
Unfortunatly no. The only reference in the whole manual is: "AC: block all the DC components and attenuate signals lower than 8 Hz. Use AC coupling to get a stable edge trigger when your waveform has a large DC offset."
Of course I can move the trace up and down (it is called offet too), but that doesn't fix the issue. The trace mean value is over the offset level.
RMS can be measured for AC only, or for AC+DC. In my case, it takes into account DC (and a wrong DC!) in the AC mode, which is not as expected.
No. I can set a 1 s/div and take a long capture, and the trace mean value is always above.
KeithWalker
Joined Jul 10, 2017
3,092
If you can move the trace up and down in AC mode then I don't understand your problem. Does the mean value shown change with the position of the trace?
Elerion
Joined Sep 11, 2017
125
kaindub
Joined Oct 28, 2019
132
The signal is AC coupled. Even if its AC coupled there will be a DC offset if the area under the positive peaks does not equal the area under the negative peaks (its called integration).
I believe what you are seeing is correct
Prove it yourself by coupling in a clsoe to perfect sine wave and see if the offset still exists.
Elerion
Joined Sep 11, 2017
125
Yes, there is. If I disconnect the probe (nothing connected to the BNC input), there's an offset too. If I set the time base to something like 10 or 20 ms (that's already pretty slow), there's still a non-zero mean value.
I updated firmware and did a self callibration successfully. Still, the offset is there.
Could it be a fault? Sometimes is larger, sometimes is lower (same setup).
Today, while working on a real circuit, I got zero offset for a couple of minutes. Then, the offset came back (same circuit, same probe, same attenuation, etc.)
Any idea?