Hi, why do I keep getting a voltage of 60mV across the inductor in the following series RL circuit:

Resitor: 2.2k
Inductor: 8.2mH
AC source: 13 volts at 0 degrees

The voltage across inductor is supposed to be 18.27mV at an angle of 89.92, with a circuit current of 5.91mA at -0.08051 deg.
However, when I connect my 10meg digital voltmeter I keeep getting about 60mV.

-----------------------
I have a theory, but I'm not sure. I think the voltage might be unmeasurable with voltmeter due to the fact that it is 89.92 degrees ahead of the current....
Also what I´m probably getting is the current multiplied by 10: 5.91 x 10 = 59.1mA, which is what the voltmeter is probably picking up as voltage across inductor.

---Please, any help.
 

t_n_k

Joined Mar 6, 2009
5,455
Looks like the source frequency is 60Hz - based on the OP's calcs.

I think Ghar has it. The inductor will have a resistance value which the OP has not included with the calculation.

I have a theory, but I'm not sure. I think the voltage might be unmeasurable with voltmeter due to the fact that it is 89.92 degrees ahead of the current....
Also what I´m probably getting is the current multiplied by 10: 5.91 x 10 = 59.1mA, which is what the voltmeter is probably picking up as voltage across inductor.
victorhugo289,

Your theory unfortunately doesn't make much sense. The voltmeter will simply respond to the voltage magnitude across the test leads. On AC range a DVM is usually calibrated for RMS value indication of a sinusoidal waveform - unless it's a true RMS responding meter, in which case the waveform doesn't matter. The meter won't respond to or be subject to the phase relationships in the circuit.
 

Thread Starter

victorhugo289

Joined Aug 24, 2010
49
Hi all, thanks for replying.
--The frequency is 60Hz. (It comes from the secondary winding tap of a 24 volt transformer connected to 120 v) Reactance is about 3.08 ohm.
As to how accurate the inductor value might be, well I don´t know, I took it from an old CRT monitor, it explicitly labeled "8.2mH" which was a surprise to me, it is the size of half an inch but it has very fine wire wrapped around.
I hope this helps, and you can tell me why I keep getting the 60mV.
My voltmeter is not true RMS, it is a normal voltmeter.

As a side note, I've made a test with another resistor and I think I got too the value of current multiplied by 10 as in the first circuit!
 

Georacer

Joined Nov 25, 2009
5,182
Use your voltmeter to measure the resistance of the inductor. Do this just as you would have done in a usual resistor. Just take the measurement at the two endpoints of the wire, but wait a little for the indication to stabilize, as the inductor will resist the measurement at first.

Tell us what you measured.
 

t_n_k

Joined Mar 6, 2009
5,455
As a side note, I've made a test with another resistor and I think I got too the value of current multiplied by 10 as in the first circuit!
Does your meter have AC current range?

If not, how are you attempting to measure the current? It would be a very unusual meter if you had to divide / multiply indicated readings to obtain the true current value.

Perhaps you might consider submitting a schematic diagram of how you are doing the experiment.

I'm also puzzled by this statement ...

--The frequency is 60Hz. (It comes from the secondary winding tap of a 24 volt transformer connected to 120 v)
You state that you are using a 120:24V step down transformer as your supply (from secondary) but only have 13V actually measured as your source voltage?
As stated in post#1...

AC source: 13 volts at 0 degrees
The 11V difference in expected and measured (?) secondary voltage is a significant discrepancy.
 