Transient Response Calibration
Discussion in 'General Electronics Chat' started by Ryne Shelton, Feb 20, 2015.
  1. Ryne Shelton
    Feb 20, 2015
    I am responsible for calibrating the transient response for a DC Power Supply. I have no idea how to measure this. Any thoughts?
  2. MikeML
    Oct 2, 2009
    5,451
    1,066
    Use a big NFET to connect/disconnect a several amp resistive load to the power supply. Measure the step response of the PS as the NFET turns on/off.
    This illustrates the concept. Note how bad it is without a bypass cap on the output:
  3. Ryne Shelton
    Feb 20, 2015
    Okay, I'm not sure if we have any NFET's here in the lab. Any alternate methods that you can think of?
  4. MikeML
    Oct 2, 2009
    5,451
    1,066
    What kind of lab are in if you have no NFETs?
    How about a relay? It would likely have contact bounce on make, but should have a clean break.
  5. Ryne Shelton
    Feb 20, 2015
    I'm in a calibration lab. I'm also new to the lab so I'm not familiar with everything in it yet. Everyone else is gone.
    I'll look around and see what I can come up with. Thank you for the help.
  6. crutschow
    Mar 14, 2008
    13,522
    3,390
    Even a toggle switch should work to connect and disconnect the load.
    The contact bounce on make should be tolerable for this test.
  7. MikeML
    Oct 2, 2009
    5,451
    1,066
    Would work if the calibration lab has a single-event recording oscilloscope or data acquisition system.
  8. Ryne Shelton
    Feb 20, 2015
    Thanks for the info guys. And yes, we do have single-event recording o'scopes. I'll try it on Monday when everyone is back in the lab and let you guys know how it goes. Have a good weekend.