Rigol DS1052E Oscilloscope acting wierd

Discussion in 'General Electronics Chat' started by LarsHansen, Aug 30, 2016.

  1. LarsHansen

    Thread Starter New Member

    Mar 3, 2014
    2
    0
    Im rather new to electronics, and Im really in trouble here.

    Im have a bit of a puzzle here, regarding my Rigol DS1052E Oscilloscope. Im trying to adjust the probes but get a wierd result on ch 1. For both channels I use the same settings: Coupling = DC Bandwith limit = on Probe = 10x (Also 10x on the actual probe) Digital filter = OFF.

    I use the squarewave on the scope as input for both channels, and I use the same (one) probe for both channels. Channel 2 gives me the expected squarewave with 3 volt signal. Channel 1 however gives me 1/5 = 600 mV and the signal is noway near a squarewave. Se attached photos.

    Please give me your advice. If I can assist you with further information, then please feel free to comment regarding this.

    Thanks in advance.
    20160830_212126[1].jpg 20160830_215207[1].jpg
     
  2. AlbertHall

    Well-Known Member

    Jun 4, 2014
    1,911
    379
    Compare the channels with AC coupling selected.
    Channel 2, same square wave, different DC level.
    Channel 1 ??
     
  3. BR-549

    Well-Known Member

    Sep 22, 2013
    1,981
    388
    I am not familiar with your scope. By looking at the photos, it appears to me that you don't have the proper settings for a comparison. I see a couple of setup differences.

    In the first photo, you only have an input at cur-b. But in the second photo, you have both inputs. Also, in both photos, bottom right, cur-b is selected, if that means anything.
     
  4. LarsHansen

    Thread Starter New Member

    Mar 3, 2014
    2
    0
    cur-a and cur-b are cursers which are used to measure between two points. It has no relevance to the problem.
     
  5. BR-549

    Well-Known Member

    Sep 22, 2013
    1,981
    388
    Ok, why does the cur at top of screen in top photo show an input on cur-b and no input at cur-a, but in bottom photo shows two inputs?

    Just curious. I have been tempted to buy one.
     
  6. atferrari

    AAC Fanatic!

    Jan 6, 2004
    2,648
    761
    I got a different brand but I can suppose that yours has it explained in the User Manual. Isn't it?
     
  7. crutschow

    Expert

    Mar 14, 2008
    13,000
    3,229
    If you are just moving the probe from input A to input B without changing anything else then it would seem you have a problem with your oscilloscope.
    Note: If you take pictures without the flash the screen will show up much better.
     
  8. SLK001

    Well-Known Member

    Nov 29, 2011
    812
    225
    Is channel 1 AC coupled?

    EDIT: I see that it is DC coupled. What did you measure on chan 1 that could have damaged the input?
     
  9. Lestraveled

    Well-Known Member

    May 19, 2014
    1,957
    1,215
    It appears to me that you have a problem with channel one.

    Question - Does this symptom happen on all scale settings?

    Regardless, your scope needs repair. Please keep us informed.
     
  10. AlbertHall

    Well-Known Member

    Jun 4, 2014
    1,911
    379
    I suspect that this problem is between the input socket and the Y-amplifier - Coupling switch/attenuator/compensation - possibly damaged by overload but other causes are possible.
    What do you get on channel 1 if you use AC coupling? If it is significantly different when using AC coupling then the problem is in the coupling switch circuitry.

    Do you plan on repairing this yourself or just confirming that the scope is faulty (it is)?
     