pH Glass Electrode
Discussion in 'Homework Help' started by lam58, Dec 28, 2015.
  1. lam58
    Jan 3, 2014
    69
    Hi, i'm stuck on part b) of the question below:
    Q: In order to perform pH determinations with a glass electrode, the cell potential was measured for threestandard solutions with the following pH values at 25 Celsius: 2.04, 7.05, and 9.20. The cell voltage readout(in mV) for each of the above solutions was 238.0, -37.5 and, -164.5, respectively. Calculate: (a) thesensitivity of the pH sensor; (b) the pH of an unknown sample yielding a cell voltage of 20.5 mV; (c) thepH deviation from the actual value if the sample temperature is 35 Celsius.
    My ans:
    So to get the sensitivity (s) I just use the nernst equation i.e.
     s = \frac{2.303 RT}{nF}
    Where: R is the gas constant     8.3144 J.K^{-1}.mol^{-1}
    T is temperature in Kelvins
    n is the charge of a Hydrogen ion = 1
    F is Faraday's constant   96485 C.mol^{-1}
    \Rightarrow \frac{2.303 \times 8.3144 \times 298}{96485} = 0.0591 V/pH unit
    But now I'm stuck on part b). My method is either to rearrange the nernst equation like so:
     E = E^0 + \frac{2.303 RT}{nF}. pH
    \Rightarrow pH = \frac{E - E^0}{2.303 RT/ nF}
    where E^0 = 238mV - 37.5mV - 164.5mV = 36mV
     = \frac{0.0205v +  0.036v}{0.0591 v/pH unit}
    But this gives me a pH value of 0.96 which doesn't seem right.
    The other method I am using is to work out the value given the pH scale and sensitivity. i.e.
    If we know that the cell voltage of the unknown pH is 0.0205V, then if we find the difference between this and the the voltage for pH 7.05 (0.0375V) gives 0.058V. Then divide by the sensitivity of 0.0591V/pH unit to give a pH difference of 0.981. Then then take this value away from the reference pH value of 7.05 to give a pH of 6.068.
    Any thoughts on my methods, are they correct or am I way off?
  2. sailorjoe
    Member
    Jun 4, 2013
    361
    63
    It's rare to see a question about electrochemistry on this site, but hey, it's all just physics.
    I'm no expert, and I think there is a problem with the the text of your numbers in the post, but my first instinct is for question a), just show the relationship between the cell potential and the pH values. It's a graph of one vs the other, just to show how much one varies with respect to the other. Question b) then becomes a lookup on the graph. And question c) is where you apply the Nernst formula to compensate for temp change.
    Others may disagree, and I would defer to more experienced chemists.
  3. bertus
    Administrator
    Apr 5, 2008
    15,799
    2,386
