Oscilloscope Probes?

Discussion in 'General Electronics Chat' started by ESSO, Mar 13, 2015.

  1. ESSO

    Thread Starter New Member

    Feb 14, 2015
    14
    0
    Hi everyone!

    Last week in class, we were using an oscilloscope in a lab for Op Amp, well we have 2 channels, 1 - for input and 1 - for output; The probes were different colors, the channels were setup the same (Probe 1x) However Probe 2 (output ) was not given any gain, actually it was less, only until we change the Probe 2 to 10X (in the Oscilloscope) is when we were able to see the gain to what we calculated. After changing the probe to the same color, we were able to see our gain, with the 2 channels setup equally setup to 1x. I assume the probes different colors have different setup? Is this true? or we had the Oscillos setup wrong - (Teacher review setup and everything was ok) ? I read in the forums about impedance and frequency, Is this had something to do with our reading?

    Thanks for sharing

    Gera
     
  2. alfacliff

    Well-Known Member

    Dec 13, 2013
    2,449
    428
    did anyone check the adjustment on the probe? most scopes have a square wave output to calabrate the probe, and a variable capacitor on the probe to adjust it.
     
  3. flat5

    Active Member

    Nov 13, 2008
    403
    17
  4. ESSO

    Thread Starter New Member

    Feb 14, 2015
    14
    0
    Both Probes have a buttom, 1x-10x, both were set at 1X, Did not know about the other adjustment and I believe neither the instructor, since he did not mention anything
     
  5. ESSO

    Thread Starter New Member

    Feb 14, 2015
    14
    0
    Thanks Flat5, I read it, there is more info then what i know at this moment... But as soon as I go back to school (Spring B), I will look into more.
    It appears to me that one is Passive and the Other one is Active. _ But That is my guess..
     
  6. alfacliff

    Well-Known Member

    Dec 13, 2013
    2,449
    428
    x1 probes are passive. also x10 probes with passive attenuatores. there is usually a small screwdriver slot adjustment to compensate the internal LC of the probe to the osciliscope. adjust for best square wave when hooked to "cal" on scope.
     
  7. Roderick Young

    Member

    Feb 22, 2015
    408
    168
    It does sound like the second probe that did not show gain was a 10x probe, or was an adjustable probe set to 10x, meaning it would deliver only 1/10 of the signal to the oscilloscope. If your readings were off by exactly a factor of 10, that would strongly suggest that this was the case.

    My apologies for the the slight content drift, but I would strongly suggest that you continue to pursue the matter until you understand exactly what the issue was. More than 30 years ago, one of my first managers advised me that if there is something that doesn't seem quite right in an experiment, always look into the root cause; that is, don't ignore a problem that seems to have vanished without knowing why, because that little problem will come back later and bite you in the *** (I'm censoring his exact word). That character trait of not letting go until one fully understands seems to be the mark of a good engineer.
     
    RamaD and ESSO like this.
  8. Reloadron

    Active Member

    Jan 15, 2015
    964
    232
    When it comes to scope probes, some probes are 1X, some probes are 10X and some probes are 100X. Some probes have a switch to select 1X or 10X. So a 1X probe offers no attenuation to the incoming signal, a 10X probe attenuates the signal by a factor of 10. Some probes have the ability to communicate what they are to the scope and the scope will range the vertical input gain accordingly and some probes do not have this ability and nor does the scope they are connected to. With a focus on only passive attenuator probes. Oh wait, some probes that are 1X or 10X even have a Ground center position which grounds the scope's vertical amplifiers, it does not ground the probe tip.

    When using a 10X or 100X probe the probe needs to be calibrated to the scope vertical channel it is used with for frequency response. If we look at the vertical input channels on most scopes they tell us the input impedance generally 1 M Ohm and they also show a capacitance which shunts the vertical input channel. Something like 1 M Ohm / 25 pF. The probe needs to be compensated to the vertical input channel. Alfacliff mentions this. Most scopes have a signal that can be used for probe compensation. Generally a 1 KHz square wave around 1 or 5 volts in amplitude. The probe has small holes allowing for adjustment.

    A few examples:

    Under Comp.png

    High Frequency Attenuation, note how the leading edge of the square wave is attenuated. The probe is not properly matched or compensated to the vertical input channel.

    Over Comp.png

    High Frequency Accentuation. The accent is on the high frequencies of the waveform.

    Normal Comp.png

    Just right where the square wave corners are nice and sharp. This is normal compensation.

    Everything about scopes and passive probes is a matter of the scope used and the probe used.

    Ron
     
    ESSO likes this.
  9. ESSO

    Thread Starter New Member

    Feb 14, 2015
    14
    0
    Good stuff!, Definately I will go back next week and look into it more, At this moment I dont have the model of the Oscilloscope.. or the probes. I am not sure if school ever send the Scope to Calibration or instructors do it.. Thanks guys.. I used a scope 15 years ago in school and never used it again in the industry that i am in .. Now that I went back for my B.S. in EE is where I am starting to play again :)
     