****************** SOLVED THE PROBLEM *********************
I am fixing a Fluke 8840A's PS output +7.5V (TP604).
I changed the 7.5V Zener diode (CR613) 12 times from 3 different suppliers and from two other MBs. Always I got around +4V (3.8~4.5) never over 5v.
I put new resistor (R601) and capacitor (C609).
I always checked the part before I put in and after I pulled out to see if I put too much soldering iron heat and damaged.
All the voltages out of TP601 through TP607 has no ripple.
At last I pulled out the the resistor (R601) and the Zener from the MB. Put the two parts on a BB and connected my Lab PS to 15V. My DMM shows 7.69V across the Zener!
No clue what I have to now.
Any suggestion will be big help, sure!
I am fixing a Fluke 8840A's PS output +7.5V (TP604).
I changed the 7.5V Zener diode (CR613) 12 times from 3 different suppliers and from two other MBs. Always I got around +4V (3.8~4.5) never over 5v.
I put new resistor (R601) and capacitor (C609).
I always checked the part before I put in and after I pulled out to see if I put too much soldering iron heat and damaged.
All the voltages out of TP601 through TP607 has no ripple.
At last I pulled out the the resistor (R601) and the Zener from the MB. Put the two parts on a BB and connected my Lab PS to 15V. My DMM shows 7.69V across the Zener!
No clue what I have to now.
Any suggestion will be big help, sure!