How to measure small change in the intensity of laser (red - near IR) light after absorption of gas?

Discussion in 'General Electronics Chat' started by viswapravin, Mar 16, 2016.

  1. viswapravin

    Thread Starter New Member

    Mar 16, 2016
    21
    0
    Using photodiodes/ phototransistors are effective for this measurement, as the absorption spectrum traces are in the range of 10^-24.
    Also how to choose the photodiodes/phototransistors
     
  2. jpanhalt

    AAC Fanatic!

    Jan 18, 2008
    5,685
    900
    10^-24 of what? Please add units if they exist, or if not, then what is it a count or ratio of? If you already know what devices will work, please give an example of that device.

    John
     
  3. Alec_t

    AAC Fanatic!

    Sep 17, 2013
    5,791
    1,103
    How small is a "small change in the intensity"?
     
  4. viswapravin

    Thread Starter New Member

    Mar 16, 2016
    21
    0
    I read the absorption line intensity of the order of 10^-24 cm/molecule after absorption. The laser power is 10mW. Can I use photo diode/transistor to measure the change in intensity. If so, then what will be the range of voltage difference.
     
  5. viswapravin

    Thread Starter New Member

    Mar 16, 2016
    21
    0
    I read the absorption line intensity of the order of 10^-24 cm/molecule after absorption. The input power of laser is 10mW. Can I use photo diode/transistor to measure the change in intensity. If so, then what will be the range of voltage difference.
     
  6. Sensacell

    Well-Known Member

    Jun 19, 2012
    1,129
    266
    Convert this "small change" to an AC signal.

    Then you can process this signal with simple electronics.

    I once worked for a company that made IR gas analyzers using rotating IR filters on a wheel...
     
  7. viswapravin

    Thread Starter New Member

    Mar 16, 2016
    21
    0
    can you please explain..?
     
  8. Sensacell

    Well-Known Member

    Jun 19, 2012
    1,129
    266
    It's was all a bit complex...

    They used a black body IR source (broad spectral output) with a rotating filter wheel in front of it.
    The filter wheel had IR band pass filters that would select a specific wavelength of IR light to pass through the sample chamber.

    One wavelength was chosen to be right on the absorption band of the gas in question, the other was a reference wavelength that was not absorbed. The output of the detector was an AC signal that was amplified and fed into an ADC, along with a sync pulse from the rotating wheel.

    By comparing the "reference" and the "absorption" signals, you could determine the concentration of gas, even when the sample chamber windows got dirty. The wheel had filters for lots of different gasses.
     
  9. KeepItSimpleStupid

    Well-Known Member

    Mar 4, 2014
    1,144
    203
  10. wayneh

    Expert

    Sep 9, 2010
    12,103
    3,038
    It's also a common technique in to split the beam, and pass only part of it thru the sample. This allows precise zeroing when the sample is not present, and corrects for any fluctuation in the sampled intensity due to changes in the beam intensity.
     