Have a debate at work about analog meters

Discussion in 'General Electronics Chat' started by bwd111, Jul 25, 2013.

  1. bwd111

    Thread Starter Member

    Jul 24, 2013
    117
    1
    Meant we are having a Debate at work... The debate is on analong and ammeters that read either avergae ,peak to peak ,maxium voltage. I say Average voltage on an analog meter
     
  2. crutschow

    Expert

    Mar 14, 2008
    13,014
    3,234
    An analog meter with no electronics typically reads the average voltage with the readout calibrated to read the value in RMS for a pure sine-wave. Some old RF analog meters detect the peak voltage with a rectifier, also calibrated to read the RMS value for a pure sine-wave.
     
    bwd111 likes this.
  3. bountyhunter

    Well-Known Member

    Sep 7, 2009
    2,498
    507
    what they read depends on the waveform being measured.
     
  4. MrChips

    Moderator

    Oct 2, 2009
    12,442
    3,361
    There is such a thing as a hot wire ammeter that measures true RMS value independent of the shape of the waveform.
     
  5. bwd111

    Thread Starter Member

    Jul 24, 2013
    117
    1
    I was told its called effective which is about 70% of the peak
     
  6. MrChips

    Moderator

    Oct 2, 2009
    12,442
    3,361
    What was the debate?
     
  7. bountyhunter

    Well-Known Member

    Sep 7, 2009
    2,498
    507
    Like I said, it depends on the waveform. RMS (sometimes called effective) is .707 of peak only for a sine wave.
     
  8. The Electrician

    AAC Fanatic!

    Oct 9, 2007
    2,281
    326
    The RMS value of a waveform is equal to .707107 times the peak value for a very large number of waveforms, not just sine waves.

    Here are two easy to find example waveforms:

    [​IMG]
     
  9. The Electrician

    AAC Fanatic!

    Oct 9, 2007
    2,281
    326
    The D'Arsonval type movement responds to average value. This is the type of movement that the traditional so-called VOM uses.

    The moving iron (iron vane), hot wire, thermocouple, electrodynamometer and electrostatic movements respond to the RMS value.
     
  10. crutschow

    Expert

    Mar 14, 2008
    13,014
    3,234
    True. But you have to cherry-pick the waveform for that to be valid.
     
  11. WBahn

    Moderator

    Mar 31, 2012
    17,747
    4,796
    To be a bit picky, it is not ONLY for a sine wave; it is easy to construct other waveforms for which it is also true. But the sentiment is valid in the sense that if the waveform you are measuring with an average responding meter calibrated for RMS is not a sine wave, then it is all but guaranteed that the measurement will be wrong.

    Pretty much only in the sense that there are a very large number of rational values between 0.5 and 0.6. You can construct an infinite number of either.

    Off the top of my head, though, I can't think of any other "simple" waveforms for which it is true -- though there is something nagging the back of my brain whispering that there IS one. Note that part of what I mean by "simple" is where the only parameter that describes the waveform, other than the name, is the amplitude or at least where the RMS value is Vpk/sqrt(2) regardless of the other parameters. Also note that fixing the waveform according to a magic parameter, such as defining the waveform to be a triangle wave with truncated at a specific fraction of the normal peak value, is cheating.

    I just remembered the "simple" waveform for which the RMS value is Vpk/sqrt(2). A square wave that varies between 0V and Vpk.
     
  12. The Electrician

    AAC Fanatic!

    Oct 9, 2007
    2,281
    326
    There are an infinite number of such waveforms. It's the sine wave that has been cherry-picked.
     
  13. The Electrician

    AAC Fanatic!

    Oct 9, 2007
    2,281
    326
    One of the most measured waveforms is the grid voltage. It is no longer very close to a sine due to all the capacitor input power supplies in the world. If one wants an accurate RMS value, using an "average responding" meter is the wrong way to go; to do so all but guarantees a wrong measurement. Fortunately, so-called "True RMS" meters are commonly available.



    What do you mean, only? However you define "infinite", denumerably infinite is still a big number.

    You seem to be trying to find a way to rule out anything but a sine. The only reason the sine function is "simple" is that it has been defined by us humans with a single word. It is defined as an infinite series, is differentiable to all orders, retains the same "shape" after repeated differentiation; it is "magic" in a number of ways. Is there not a "magic parameter" involved in the infinite series definition of a sine, the manner in which the terms of the infinite series decrease?

    To give examples of waveforms with the property under discussion necessitates combinations of various mathematical functions, which gives a result not previously defined with a name consisting of a single word. I guess that makes these waveforms not "simple".

    I can find an infinite series to represent the clipped triangle wave; then there won't be an apparent "magic parameter".

    I would claim that even your example:

    has a "magic parameter", namely the requirement of a 50% duty cycle.

    Maybe what bothers you about my previous examples is the straight line truncation (clipping) of the underlying waveform, leading to a discontinuous first derivative. Here's a "smooth" waveform with the requisite RMS = Vpk/Sqrt(2) property:

    [​IMG]

    Alternate half waves can be flipped in polarity, giving a wavefom with a DC component like your square wave example:

    [​IMG]
     
  14. studiot

    AAC Fanatic!

    Nov 9, 2007
    5,005
    513
    Do you really mean ammeters?

    I ask because most current responding meters are atually rather more sensitive and need (internal) shunts to indicate amps.
     
  15. crutschow

    Expert

    Mar 14, 2008
    13,014
    3,234
    By cherry pick I mean you are picking waveforms that happen to have the same average to RMS value as a sine-wave. In general, that's not true, and unless you know the exact wave-shape when you measure the voltage/current you don't know whether the meter is indicating true RMS. So having a large (or infinite) number of waveforms that have the same average as RMS value is generally useless information in real-world measurements with an average responding meter. :rolleyes:
     
  16. MrChips

    Moderator

    Oct 2, 2009
    12,442
    3,361
    I believe we are all saying the same thing.
    A meter is either showing true RMS value or it is not true RMS value.

    Any conversion from average or peak to RMS is a crock.
     
  17. crutschow

    Expert

    Mar 14, 2008
    13,014
    3,234
    So you are saying we should never use an average responding meter that's calibrated to read the RMS of a sine-wave? :confused:
     
  18. The Electrician

    AAC Fanatic!

    Oct 9, 2007
    2,281
    326
    The waveforms I'm describing don't have the same average as RMS; they have the same crest factor, the peak to RMS value.

    The described abundance was cited only as a counterexample to the statement that only the sine wave has that crest factor. It wasn't intended to be an aid to measurement. It means that encountering a waveform with that crest factor (or close to it) that isn't a sine shouldn't come as a total surprise.

    But absolutely pure sine waves are a rarity. In particular, the grid voltage isn't a very good sine wave these days. If a person wants to measure the RMS value of a waveform, use a "True RMS" meter; there's some helpful information for you.

    Anybody who has a look at their local grid waveform with a scope will see how it has become flattened on top due to the large number of capacitor input power supplies on line.

    True RMS meters are becoming the rule rather than the exception. Even low cost DMMs can be had with True RMS capability.

    There is even a DMM that can switch modes, and measure waveforms either with an average responding algorithm or True RMS:

    http://www.tequipment.net/YokogawaTY720.html

    So, if you want to measure the RMS value of a waveform, get a True RMS meter. :)
     
  19. crutschow

    Expert

    Mar 14, 2008
    13,014
    3,234
    Average responding meters respond to the average of the waveform voltage. Why are you talking about the peak to RMS crest factor which is only pertinent if you have a peak responding meter (which are rare), not an average responding one? :confused:
     
  20. MrChips

    Moderator

    Oct 2, 2009
    12,442
    3,361
    What we are saying is get a True RMS reading meter and there is nothing to argue.
     