Circuit explanation

studiot

Joined Nov 9, 2007
4,998
What is your interest and background?

IMHO persons should not be promoting circuits connecting across the human heart without knowing what they are doing.
 

Thread Starter

kavin6

Joined Jan 2, 2009
16
What is your interest and background?

IMHO persons should not be promoting circuits connecting across the human heart without knowing what they are doing.
To design an ECG amplifier. I have done it and its working fine but I need to know the working of it.
 

ifixit

Joined Nov 20, 2008
652
I have had some design experience with instrumentation amplifiers (INA) connected to patients many years ago. I doubt the design requirements have been relaxed much. The main requirement is to measure a bio signal from within the body, while at the same time reject all the extraneous electrical noise that is around us at all times. You are not allowed to harm the patient either.

Signal common mode noise is always present and must be kept to a minimum to be within the operating range of the INA118. Taking a sample of the center of the gain resistor (Rg) gives you the common mode signal of LA and RA, which is noise and 60Hz hum from around the patient. This signal is buffered to get a low impedance output used to drive the shield and "guard" the input from stray leakage currents.

The guard signal is then inverted, amplified, and converted to a current, which is "injected" into the patient’s leg. This has the effect of canceling the unwanted noise on both RA and LA equally. This connection also provides a safe virtual ground reference to the patient. I recently had an ECG myself and both legs had a connection, which I suspect distributes the feedback signal more evenly.

Connecting electrical equipment to a human patient is a situation which scrutinized very carefully by the electrical safety authorities and a lot of safety requirements need to be met. Even just connecting a grounding electrode to a patient can be a problem if that ground is used by, or effected by, other equipment in the building (e.g. a hospital).

Please be very careful with what you are doing.

Regards,Ifixit
 

Thread Starter

kavin6

Joined Jan 2, 2009
16
I have had some design experience with instrumentation amplifiers (INA) connected to patients many years ago. I doubt the design requirements have been relaxed much. The main requirement is to measure a bio signal from within the body, while at the same time reject all the extraneous electrical noise that is around us at all times. You are not allowed to harm the patient either.

Signal common mode noise is always present and must be kept to a minimum to be within the operating range of the INA118. Taking a sample of the center of the gain resistor (Rg) gives you the common mode signal of LA and RA, which is noise and 60Hz hum from around the patient. This signal is buffered to get a low impedance output used to drive the shield and "guard" the input from stray leakage currents.

The guard signal is then inverted, amplified, and converted to a current, which is "injected" into the patient’s leg. This has the effect of canceling the unwanted noise on both RA and LA equally. This connection also provides a safe virtual ground reference to the patient. I recently had an ECG myself and both legs had a connection, which I suspect distributes the feedback signal more evenly.

Connecting electrical equipment to a human patient is a situation which scrutinized very carefully by the electrical safety authorities and a lot of safety requirements need to be met. Even just connecting a grounding electrode to a patient can be a problem if that ground is used by, or effected by, other equipment in the building (e.g. a hospital).

Please be very careful with what you are doing.

Regards,Ifixit
Thanks a lot ifixit :D
 